What happens if you just turn Bluetooth off on your phone? Do you still get those problems? Or are you needing to make adjustments via the phone during the day? (No streaming phone calls but a quick error check)
One thing obviously different when you were on vacation is that you didn’t use the wifi router at home. I think you said you switched to 5 GHz and it seems to have helped. But it didn’t solve the random reboot issue altogether? If not, there’s a possibility that your router is generating some kind of interference (maybe with or without the TV Adapter) that causes the OPN S1 to reboot.
The issue you have with connecting to the ON app on your new iPhone is a different issue (connection issue) so if you had the new iPhone with you on vacation and it didn’t reboot then it just established that it’s a different issue and not inter-related.
The other factor is the TV Adapter 3.0 which you didn’t have with you on vacation. You did say that you had experienced reboot when or when not using the TV Adapter before. But one difference is that when you’re at home at didn’t use the TV Adapter and still had reboots, maybe the TV Adapter was connected to power and maybe even turned on, even if you didn’t put your OPN S1 in the TV Program mode. So one thing to try is to make sure the TV Adapter is removed from power when you’re not using it.
Thanks… the only problems are: 1) I didn’t have any of these issues when I had previous iPhone X - I’ve had the same TV box same Wifi router, same TV ,etc… 2) problem only started with new iPhone 12 Pro … seems to me it’s something with the combination of the new Iphone 12 Pro and TV box , and the app…
One new difference on the iPhone 12 is that it is 5G capable. I wonder if this causes some kind of interference or not. If it’s possible to disable the 5G capability on your iPhone 12 to just use 4G or LTE like before, then that would be worth a try.
I did try using the non-5g and it didn’t really help… not sure how to “disable” it?
My step son will be here tomorrow and he has the iPhone 12 , I will ask him if I can try his phone with my OPNS1 aids. I have a new WiFi system, and all types of WiFi devices enabled all at home once, him here, there will be at least 18 WiFi devices enabled and connected all at the same time. Now I have to say that my router s setup to force 5gigs when ever possible, but I do have a few 2.4gig that moly devices.
Will be interesting to see what happens… Thanks
Ok I went to audiologist today and she’s sending the aids to Oticon for a check. I got loaners ( Opn1. She had OPNS1 but they wouldn’t charge.
Bad news is same issue with aids randomly rebooting while connected to TV box so I guess it’s the TV box. My aids are still under warranty so shouldn’t cost anything for the check but guess I have to look for new TV box.
Still weird about aids rebooting and I don’t understand why TV box is causing the issue but.
Just a quick update… I went to Audiologist today… Oticon sent a new pair of OPN S1s for me… I asked if they told her what was wrong with my other ones… She said they didn’t say anything to her… just sent a new pair… So, I’ll see what happens with TV box streaming… she did order a new TV box, but we decided I’d just use the TV box I have to see if they problem of the aids restarting continues… I did take my TV box and she checked software and confirmed that my TV box has most current firmware (2.1.1 according to her & Oticon…)
UPDATE… well the new S1s are kind of doing the same thing when connected to TV box They aren’t “restarting” but one or the other of them will disconnect from TV box, and then reconnect to the box… the other thing is that TV box volume is back to being LOUD … not sure why that is, as audiologist said she programmed the new aids based on what I had in her software… any ideas on this? I may have to ask audiologist to borrow the new TV box she got (probably should have asked to borrow it today…) and see if the issue persists with a different TV box…
The only thing I can think of is that the volume setting for it in Genie 2 is set to more than the default. See the TV Adapter/EduMic Loudness setting in the screenshot below. Maybe instead of being set at the “down arrow” level, it’s set to the right or rightmost level.
Did you simply pair your new OPN S1s with your old TV Adapter 3.0 on top of the old OPN S1s? Meaning that your TV Adapter is now paired with both the old and the new set? Before trying out your audi’s new TV Adapter box, it may be worth trying to perform a system reset on your current TV Adapter box to clear out the pairing of the old set to see if that helps with the disconnect/reconnect flakiness you’re seeing or not.
Thanks… Re TV box volume: before, when I would connect to the TV box, via the app, the volume would show up in the app as “14” … it was still too loud and I’d just adjust it … now, it’s defaulting to “16” and again, I just have to turn it down… the audiologist has adjusted it several times in the past but for some reason, it never really “sticks” … I think she’s put it all the way it second to lowest setting… Since I got the new S1s today - would the setting she had for TV adapter been transferred over to the new ones? I’m assuming so?
She paired the new aids to the TV box since I brought it with me … I’ll try the reset of the TV box and see if that makes a difference…
I’m just guessing that if you paired the OPN S1 first to the hearing box, maybe that’s when the TV Adapter “takes” the setting for the TV Adapter box from the hearing aids as was set by Genie 2, which at that time was on the louder volume setting. Then you discovered that it was too loud, asked your audi to change it to a lower volume setting, which the audi did, but then you didn’t re-pair the hearing aids (now with the new change) to the TV Adapter again, so the new Genie 2 setting made to the hearing aids didn’t make it to the TV Adapter, and maybe that’s why it didn’t stick.
This is just a guess, but if that’s what you observed, it may help to perform a system reset on the TV Adapter box then re-pair the hearing aids to it, just to clear any old settings out first to allow the new pairing to take in the new setting. At least if that still didn’t help, at least then you know the hypothesis above isn’t true.
At any rate, it doesn’t hurt to perform a system reset on the TV Adapter box and the only cost of doing that is the extra time it’d take you to have to pair your hearing aids to it again. But at least you know that you start out fresh with nothing old still lingering in the box.
The TV Adapter box allows an unlimited number of hearing aids to be paired with it, but if its firmware is not robust enough, things may get mucked up. If you’re the only one using it, it wouldn’t hurt to perform a system reset before doing a pairing of your new hearing aids to it.
Thanks… I just did system reset on TV adapter (pressing red button on the back - I assume that’s what you meant?) … re-paired the aids - still started out at 16 (in the app) … I guess I’ll email the audiologist - telling her about the disconnect/reconnect to TV in the app issue, as well as the volume… (right aid just did the disconnect/reconnect to TV program as I’m typing this) … I’ll probably ask to borrow the new TV box she got to see if that solves the problem … I’m thinking that perhaps something in the TV box is going “bad” (with Bluetooth I guess?) … assume that’s possible, unless these TV boxes are built to last “forever”
My TV adapter always starts out at 16 and I normally lower it to 15 and also set the environment down to -1. Not sure if this relates to your issue but if I have the connect clip powered on and connect to the TV adapter my ON app drops connection to the aids. This is with the new More1 aids, a new TV adapter and new connect clip. And at the time I first paired my More1 aids to the iPhone I rebooted the iPhone first.
Thanks… I don’t have Connect Clip … I did reboot the phone, deleted the app, re-added the app, re-paired the aids, rebooted the TV adapter box, etc,etc… And the aids take FOREVER to fully connect to the app… usually at least 10 minutes or so, before the aids actually fully connect … right now, it’s showing “left disconnected”… been that way for about 5 minutes now…
I think I read somewhere/sometime that TV box volume can still show 16 even though Genie software will be set at lower setting (5 settings I think…) . As I said, with previous aids (and the loaners I had for the past week), TV volume in app would show 14 - now, with new S1s, it’s back to showing 16 and definitely LOUDER than it was before… Not sure if audiologist can check the settings without me being there, but I’ll contact her Monday with this issue, as well as the aids disconnecting/connecting to TV box - as I said, at least they aren’t rebooting like the previous aids(and the loaners) , but I still want to try the new TV box to see if that solves the issue…
Do that you have the aids paired to an iPad? It sounds like that issue to me. I am also wondering if it has to do with either the WiFi or cellular signals of the iPhone. Because you said this started with getting the new iPhone if I remember correctly. You could try going to the iPhone control panel and disabling the cellular and see if changes the issues. And maybe even the WiFi too. If not either of these, maybe it is the TV Adapter but personally I don’t think it is.
Thanks… No iPad - no other Apple devices (except for Apple watch…) only thing that worked previously was when I turned off Bluetooth on the phone… that seems ridiculous if I have to do that … and, again, it didn’t happen with previous iPhone (X) and I believe the X and 12 Pro (my current one) have the exact same Bluetooth hardware ? I have tried turning off WiFi on phone and that didn’t help… haven’t tried turning off cellular but that would be silly if I have to do that … am I the only one that has trouble with the app connecting? Seems like every time I open up the app, the aids will have to go through the “connecting” dance - again, usually 5-10 minutes …
If the ON app is taking too long or never connects, the best way I’ve found to reconnect is to: kill all applications running on the phone, power the phone off, restart it (it boots fairly quickly with a 12), restart the hearing aids, the aids connect quickly (as confirmed by the MFi functions working from the newly powered up phone), and then start the ON app and it almost always connects quickly. I know that is a long procedure but it works for me. I also typically do a clean boot of my phone every day after I take it off charge - again that should NOT be necessary but it helps to make things work as one expects. It appears to me that both iOS (14.6) and the ON app have issues but I’m not an iOS programmer (but I do have programming experience on UNIX and Windows writing instrument control programs using multiple communication protocols).
As for interfering signals, I’m not sure that is the problem because the above routine fixes issues where I have a WiFi 802.11ax router (about a meter from the TV Adapter 3) and neighbors have WiFi routers all of which are running both 2.5 and 5 GHz transmitters simultaneously (I can see them using an app on my Mac) and my iPhone 12 is connected to cellular (5G signal) after the above procedure and ON works.
It is frustrating that things don’t work as they should. It appears to me that Oticon and Apple are pointing fingers and neither will fix issues. I expect neither do adequate testing before releasing new hardware (and firmware) and software versions.
Thanks…with your procedure, do the aids stay connected to the app? Whenever I open the app, the aids take 5-10 minutes to fully reconnect (usually it’s one or the other aid that shows “disconnected” and takes awhile to reconnect… I’ll give it a try…
Also, just realized that I don’t think the “start-up tones” are turned on, so when I said the aids are disconnecting/reconnecting to TV box, I think they are actually restarting (like the old ones did…)