Yeah, I should have said HCPs here and not “users” per se. You’re correct that they don’t want to promote DIY.
I also notice when I play around with adjusting the gains that the CR values may jump around and change automatically if I change the gains too much. I’m guessing that they want to make sure that the knee points on their compression chart for each frequency band are more continuous and to minimize any disjoints. Take the case of the gains and CR values at the 1 KHz example I used in one of my previous posts for the OPN S1 Music. As can be seen in the CR chart I recreated (shown again below for reference), there’s already a 10 dB disjoint at the 45 dB knee point, going from Soft to Moderate.
I would say that if you have access to force change manually both the CRs along with the gains in the Gain Controls section, you can possibly end up with very wide disjoint values at the knee points, which may introduce an undesirable effect into how smooth the amplification may sound.
(Cassin)
2. A low compression ratio of 1.7 to 1 (or a maximum of 2 to 1) is best for music.
Interestingly, if you look at my example in my previous post, the More 1 MyMusic program adheres to this rule very well, while the OPN S1 Music program has much higher CR values of as high as 3.7 in the Moderate to Loud knee points.
(Baumin article)
Purely Linear Compression Ratio
Dr. Bauman strongly recommends a pure 1 to 1 compression ratio for musicians (instead of the 1.7 to 1 ratio from Dr. Chasin’s checklist). This means that the gain of hearing aid will remain the same regardless of the input.
Hm, I guess even the experts don’t see eye to eye with each other either.
(Chassin)
3. Disable the noise reduction system for the music program.
The noise reduction system is designed to reduce the sound of the microphone inside your own hearing aid, but this feature sometimes has detrimental effects when used for music.(Chassin)
4. Disable the feedback management circuit when setting up the music program.(Chassin)
5. Disable any frequency shifting or frequency transposition for the music program.
-
Check.
-
Yes, this makes sense, especially because one of the common strategies used for feedback management is a slight 10 Hz shift, resulting in a potential fluttering effect, especially when pure tones are heard. But sometimes feedback management is a necessary evil because you can’t help it and don’t want to have feedback when trying to enjoy listening to music.
-
Yes for disabling frequency shifting used by feedback management if you can help it. In terms of frequency transposition for frequency lowering, ideally yes, you want to disable it to remove any musical aberration. But in my personal case, I keep Speech Rescue enabled even in my music program so I can enjoy some of the highs that I already can’t hear. Otherwise the music sounds more dull to me. So it’s a trade-off between having pure unadulterated music vs being able to hear something in the highs that’s missing.