I read somewhere that if you let them sit too long there is no bringing them back.
Menglxs You Da Man!!!
Took a bit of rigging but it worked.
Left them connected for only a few min, current draw was continuously dropping (About 30ua)
Oticon charger picked up and finished the charge.
Battery checks out good in Genie2, charges and lasts 20 hrs or so.
Thank you very much!!!
Hope I can return the favor sometime.
Mike
I agree with what you are saying.
I think in the case of the OPN S Rechargeable, they shut down before fully depleted and the charger will connect and charge.
I think if you leave them uncharged and they drain fully over time the charger will not pick up.
Any how the quick flash charge with 3 1.5 v (4.5v Total) in series worked for me to get it going, all good for now.
Thanks
Mike
Congratulations to you!
Please don’t think about returning a favor to me, I also got a lot of help in this forum.
Congrats on fixing your batteries. If I were you, just to be safe, I would not deplete your batteries too much or too often anymore. Meaning that even if you still have plenty of charge at the end of the day (I don’t know what it is, maybe 40 or 50%?), don’t skip charging for one night because you think there may be enough to carry you through to the next day before the battery is depleted. In short, don’t allow it to get to the point of near depletion again. And that means charging it at least every night, if not even more often, like if you decided to take an afternoon nap on a weekend or something, just flop them into the charger whenever you don’t use them.
Good advise but I normally put them in at 6am and usually have them on until 1-2 am (18-20 Hrs.)
So far so good, they have reached the shutdown point and charged right up again.
I think battery is fine, seems to likely have been left in an undercharged state for a long time and depleted even more to the point that the charger would not pick up and charge.
All good for now.
n
Thanks
Mike
Still looking for a battery source.
The battery in my Oticon OPN S is marked Z22.
The size and type is not yet recognized in Wiki.
The only Google reference I can find of the rechargeable Z22 Lithium-ion button cell battery is in a US patent…US9564654B2 - Rechargeable lithium ion button cell battery - Google Patents
The Patent says: “The capacity is found to be at 24 mAh at 5 mA discharge current, average voltage 3.7V to end voltage 2.75V, after being fully charged to 4.2V. Total energy is 89 mWh and volumetric energy density is 345 Wh/L. Remaining capacity at 80% of initial capacity is found after 600 to 1000 cycles typically at 800 cycles. Cycle regime is at 12 mA constant-current constant-voltage charge to 4.2V and discharge at 12 mA to 3.0V, 25 to 30 Celsius.”
And goes on to describe the Z22 size as diameter 7.8mm and 4.0mm high.
The inventors work at Zhuhai Zhi Li Battery Co Ltd (ZeniPower) Add: 7B, No.130, XingHua Road,
XiangZhou District, ZhuHai, GD519002, China.
I am sure one of you engineering types or other clever sleuths out there, can track down a source for this battery type.
This Oticon video shows how to change it…but not where to get it!
The battery shown as “Oticon” is the Z22 from my OPN S.
Lloyds was able to help me.
Mike
Hello!
I bought on ebay an oticon opn s1 device that is used before. And the seller did not specify that the device is rechargeable. And it came with totally dead battery. I can’t even check is it functioning. Did you solve where to buy such a z22 battery?
As I understood it’s 3.7V. Did you measure the voltage of your?
The quick answer is no. I have not been able to source an alternative source for the Z22 battery.
The good news is that you can take your aid to any Oticon audiologist/distributor and they will either change the battery there or send it to Oticon for battery change.
I must say a very big thank you to @menglxs!
I bought a pair of Oticon OPN S on Ebay some time ago. On delivery, one of them immediately charged but the other gave out the dreaded triple Red flash of death. For the last 4 months I have been looking rather forlornly at that charger and the deadly triple flash trying to source a new battery to no avail.
So as a result of this thread, I had my son fabricate a device as suggested by @menglxs. My son had a battery tray that he had previously bought from some electronics shop and used some wires, a “chip clip” (as we in Oz call a plastic clamp for resealing an open bag of crisps), two screws and a quick solder.
I was a big nervous as to what would happen while clamping up the dead Li-ion battery. But we went ahead and switched on power from the 3 batteries wired in series for 60 seconds. Then we removed the battery replaced it in the HA and then tried it in the charger. Success as the charge light came on solid Red! And after a few hours solid Green to indicate fully charged!
What a great forum this is; and how much I have learned from all of you!
Cheers and thanks.
First sign of success!
Fully charged…back in business.
Thanks for sharing this! Can you keep us up to date maybe in a month or two on whether the revived battery still continues to be charged and functional for the duration it’s supposed to last in each charging cycle? Thanks.
Congratulations, happy DIY!
I have been successful, still working normally for almost a year now.
Still looking for aftermarket replacements though.
Mike
Thanks for the update. This is great news.
I own Ryobi battery-operated tools and have seen people do similar things to revive the dead un-rechargeable Lithium-ion batteries they have. Basically if they’re drained below a threshold level then the BMS assumes that they’re dead for good and won’t allow a recharge to take place. But if you “revive” the battery to above the threshold voltage by temporarily pumping enough current into it to bring it back from the dead (by applying a higher voltage) to get the BMS to accept charging it, then you may be able to revive it.
A HUGE thank you to all of you who’ve contributed to this thread! I have Oticons that I got in the spring of 2021 and last night, Nov 1, had my right HA do the same thing! So I took a 9V battery to bring my HA battery back to life. It’s currently charging and seems fine.
One year has gone by and this battery is still going strong without problems (now with another user).
In the last few months I have also brought three different Oticon More dead batteries back to life using the same device (I am now using 30 seconds rather than 60 seconds of zapping).
They are all working perfectly again and recharging as they should…thank you again @menglxs !
Be careful Rob…4.5 volts does the job. A 9V battery might be dangerous to hit a 3.7V Li-ion battery…over-voltage could lead to thermal runaway with resulting fire/explosion risk.
And Rob, a big welcome onboard the forum!
A follow up on this topic. I had to do it a few times afterwards, I would only get 12 hours with it. Then the last time, no dice! It would not charge. I took it in to the audiologist, and she did a reset by holding the volume down button for 5 to 6 seconds, then put it directly in the charger. It charged fine, and doesn’t seem to drain as quickly. So far, so good!
I heard from a forum member (@SpudGunner) who owns the More 1, who had his rechargeable Lithium-ion battery replaced under warranty, that his audi had to reset the More 1 via Genie 2 in order for the HAs to recognize that they have now received a new set of batteries in order for it to work properly.
Originally they gave him the OPN S rechargeable batteries mistakenly as replacement for his More batteries and it was not reset (the audi was on vacation so the receptionist didn’t know better) and it was still acting weird. After the audi got back, he reset the HAs to the new batteries but it was still acting weird. Then he realized that the receptionist doled out the wrong battery for the OPN S instead.
But after the new batteries specifically designed for the More were ordered and used as replacement, and the HAs were reset again, then everything is fine again so far.
I’m guessing that the HAs keep track of the batteries’ performance history and tell the charger how to charge accordingly. So by reviving your batteries with no reset, the HAs still remember them as having a lower capacity and probably only allow the charger to charge only up to that lower capacity. By resetting the HAs, they now think the batteries are freshly new so now the HAs probably start charging them to the full capacity.