Resound Omnia 961 rechargeable versus 962 with 13 zinc-air battery

@ssa could provide a far better answer than I could, but the BT radio and any acknowledgments it has to send are far less active when there is no streaming. The HA is acting as a RECEIVER in streaming and then amplifying the radio signal it receives. BT LE means it’s not active when it doesn’t need to be and transmits only in bursts, not a continuous stream. BT Classic involves continuous ON activity. By contrast, the Phonak Marvel battery life decreases from 24 hours just listening to environmental sounds to only about 11 hours when both constantly streaming and listening to environmental sounds at the same time. Phonak unveils Lumity hearing aid platform - #73 by jim_lewis. I would imagine that BT LE signals from a smartphone to an HA are otherwise only emitted when the user changes HA settings on the smartphone.

Here’s what I means that most battery power in HA’s that are not heavily streaming is consumed just by running the onboard electronics. The following line is from a table of Omnia 961, 962 specs for LP (two left entries) and MP receivers (two right entries) for Quiescent and Operating HA’s, You can see that ~80% of HA power is just consumed “idling.” I should imagine “quiescent” is when the dang thing is powered up awaiting sound or streaming input but there is none.

Source: datasheet for standard receivers on following page: Hearing aids ReSound - ReSound OMNIA - Hearing Aids

Bing ChatGPT search for ISO definitions of “quiescent” and “operating” :

  • Quiescent conditions: The hearing threshold deviation is measured in the absence of any sound stimulus.
  • Operating conditions: The hearing threshold deviation is measured in the presence of a sound stimulus.

I like the M&RIE receivers, but you have to essentially require MP (medium power) receivers or less and have a certain shape type of loss. If you search on M&RIE receivers, you can come up with a bunch of posts on that subject.

Also, see the following post along with mine below it: ReSound Omnias with M&RIE Receivers - #31 by JT337

2 Likes

@jim_lewis thank you. Another question, as you have tried both 961 rechargeable and 962, could you please comment about the sizes of the two HA? Is the 962 significantly larger?

I didn’t find the 962 much bigger than the 961. They are both larger than the Phonak Lumity. I’m not size- or wire-conscious about my RIC hearing aids. In trying out colors, I was having a hard time deciding what color to go with; my hair is getting decidedly grayer. My audi suggested the black color would match the color of my temple tips of my eyeglasses (the part that goes over and behind the ears). So, that worked out well. From behind, it looks like I just have large temple tips on my eyeglasses.

1 Like

Thanks.

I would like to continue the discussion about the hypothesis of the battery saving via turning on flight mode.

First, the 80% of HA power being consumed in Quiescent conditions does support my argument that most of power is drained by the Bluetooth radio idling. Specifically, quiescent means no sound stimulus and therefore, no processing power. This is consistent with the common knowledge that radio components are the main consumption source in modern electronic devices.

Second, I have tried to use my 961R with and without the flight mode to see if the battery consumption is different. It is. With flight mode on HA and Bluetooth on my phone both turned off, meaning the Bluetooth of HA is in idle mode, Resound Smart 3D app showed that there was 40% battery remaining (2 green dots among 5, see photo) after 15h usage. The same test, with the HA flight mode turned on, meaning no radio at all at HA, I got 60% battery remaining (3 green dots). Of course there was no streaming as there was no connection to my phone in both tests. In spite of lacking the precise battery percentage of HA, I think it is safe to say that turning on the HA flight mode does increase the battery life.

I do some math, if 15h usage with the Bluetooth of HA in idle mode drops the battery from 100% to 40%, a full charge should last 25h in the same usage condition (there must be some optimization of Resound when the battery is low to extend the battery to 30h). With flight mode, 15h usage = 40% battery, then a full charge should last 37.5h, which is 50% more.

1 Like

I would expect the difference to be greater with Phonak aids that use Bluetooth classic instead of Bluetooth LE.

Since you only have crude charge measurements and you haven’t controlled for the sound during any listening period, your conclusions are very speculative at best.

When my rechargeable Quattros were new, I would measure charge drop from just charging until an LED came on until that LED dropped off, checking every 15 minutes in the 5 to 6 1/2 hour time range in the Smart 3D app. Brand new ReSound Quattros with the BT connection to my HA’s turned on would invariably get about 6 hours of run time between 60% and 40% charge with very carefully watching the LEDs come on during charging and go off during usage. You’re trying to find reasons to disagree with ReSound’s claim that a full charge lasts 30 hours with no streaming and 25 hours with constant streaming during that period, basically saying that claim is all hogwash, and they cheat at the end by having a special extended runtime. If you’re not getting the battery life claimed in rechargeable runtime, it’s much more likely there’s something wrong with your HA’s. Maybe you want to have them checked out and replaced under warranty if they’re not performing as they should, rather than apply unfounded theories that how your particular HA’s perform applies to all similar ReSound HA’s.

I’ve also had replacement ReSound Quattros that arrived with inferior runtime in the 60% to 40% runtime range and an overall runtime of less than 30 hours. What I found with those, carefully timing the drop in charge from 100% is the runtimes did very well in the 100% to 80% range and the 80% to 60% range but as soon as the SOC got down to about 60% the rate of runtime drop between 20% levels increased rapidly, i.e., shorter runtimes, from there on down to <10% charge, when the red charge warning comes on. So, in my experience with bad ReSound rechargeables (that got worse as I used them), ReSound does not extend runtime at low charge levels. Bad batteries show up as decreased runtimes at low states of charge, not so much above 60% SOC.

1 Like

No, I claim nothing about that 30 hours. I talked about it to find the consistency with my tests to show that flight mode turning on does increase battery life of my HA. As I have also said above, this is not precise and I do agree that they were crude charge measurements. However, the result and/or the conclusion are far from speculative, because the 50% gain of battery life in my tests is large enough to cover most of errors in measurement and setup, meaning maybe there will be not such 50% gain with precise tests, but the gain is certainly there.

About the claim 30 hours, I don’t want to argue about it just because I don’t have precise measure tools (it seems that Resound does not want to provide precise battery percentage in their app either). However, I would be skeptical about your conclusion “ReSound does not extend runtime at low charge levels” of your Quattros tests because the discharge curve of Li-Ion batteries is always steeper at low charge levels. Also, I would like to say that even if Resound does lower the performance of HA in low charge levels, it is the common practice and it is a good thing and I would not call it “cheat”.

It’s great to be able to exchange with people who want to know exactly how their HA work. Thanks again.

If running in Airplane Mode did make a big difference, why wouldn’t ReSound want to advertise it? They advertise that you get a longer battery life when you don’t stream. The HA’s, AFAIK, function perfectly well without a BT connection to my iPhone (I’ve lost the connection occasionally and haven’t noticed until I checked the Smart 3D app). I’m sure one will save some battery life, but the math on the 37.5 hr runtime in flight mode you calculate is not a “50%” increase. 7.5 hr increase / 30 hr runtime without streaming is a 25% increase in runtime over that advertised by ReSound. And if your 7.5 extra hrs were correct, 7.5 / 37.5 = 20% of available runtime, i.e., if turning on Flight Mode does completely shut off BT, then BT is only consuming 20% of the battery charge during normal operation with some level of sound input taken for granted by ISO standards.

Because the aids stop communicating with each other and their speech intelligence is reduced. Their programming is based upon the aids communicating with each other, sharing information.

Then what would be the point of entering flight mode relative to your HA’s except if it helped keep the batteries from going dead - but as long as you’re in flight mode (according to your thesis), the function of the HA’s would be seriously degraded. Entering Flight Mode would help comply with former FAA regs for NO wireless devices, but I think they’ve relaxed the rules for weak devices like HA’s, which can be kept on now, IIRC.

Starkey’s current advice:

  1. Don’t turn off your hearing aids or their wireless features on a plane. The FAA exempts devices like hearing aids and pacemakers because they don’t emit signals that might interfere with aircraft controls.

Source: Five do’s and five don’ts of traveling with hearing loss (starkey.com)

I think the bottom line is there is very little reason to use Flight Mode these days. It’s not even enabled by default in ReSound Fitting software (how about Oticon & Phonak?), it’s not required by the FAA, it doesn’t save much battery life at all compared to the long runtime of ReSound rechargeable or battery-powered HA’s, and according to @prodigyplace, it greatly interferes with normal hearing aid function.

Another weird thing about Flight Mode for my ReSound Omnia 962s is that I see no sign that it’s working. I enable it in Smart Fit (exit the program, reconnect, and confirm it shows Enabled checked for both HA’s). I leave my right HA off for simplicity. I open and close the battery door on my booted left HA 3x within 10 seconds. The left HA still shows up in the Smart 3D app on my iPhone as connected, it can still have its external mics turned on or off by the 3D app, have the mic volume changed, or the program changed, stream a podcast, and the left HA shows up in the Find My Hearing aid interface of the Smart 3D app, whereas the turned-off right HA was last seen minutes ago. The left HA shows up in an iPhone BT scanner app, too. It doesn’t matter if I fully open the battery door or only open it to the partially-open catch (where it’s normally turned off anyway).

It would be interesting to hear from other folks if the rechargeable Omnias or any other HA brands show up at all in a BT scanner when put in Flight Mode. Perhaps no one at ReSound bothered checking whether Flight Mode actually works for certain ReSound Omnias like the 962s because no one uses it anymore or has noticed that it’s gone south and reported it as a bug to ReSound? Or maybe my particular Omnias are both defective in entering flight mode?

One sophisticated possibility would be that my Omnia 962 only enters flight mode when the iPhone BT is also turned off but ReSound makes no mention of having to do anything about one’s smartphone in disabling wireless function for an Omnia. It just doesn’t seem to work for me. :roll_eyes:

I d spect that flight mode stops the Omnias transmitting, but doesn’t try to stop anything sending stuff to it.

Also don’t forget that apart from streaming, the Omnias use BT to have one aid send stuff to the other. That’s how all the focusing works. So turning BT off will substantially impact the overall performance of a pair of aids.

To permit you to legally use your hearing aids on flights that have restrictions of RF emitting device use. It is likely also a legal requirement to have them approved for sale in some countries.

That would only apply for the US. Manufacturers make products to be sold worldwide if possible. Gone are the days where they make variants for different countries.

The fact that my left Omnia in “Flight Mode” is detected by the Find My Hearing Aids” in the 3D app as last seen “just now” but the turned off right one was last seen three minutes ago indicates to me that the left HA is broadcasting and probably not in true flight mode. The signal strength meter shows a fluctuating bar for the left HA but not for the right HA, too. It’s also detected by a BT scanner app on my iPhone when in “flight mode.”

The left HA when supposedly in flight mode can also change the volume of both HA’s or the program that both HA’s are in. So, it’s transmitting to the right HA, which isn’t in flight mode. Happens even when BT on iPhone is turned off. Would be great if an HA actually sent a msg to the 3D app to confirm it was truly entering flight mode to confirm the achievement of that state for sure.

An indication that something is amiss in entering flight mode is I should hear a repeated double tone after opening and closing the battery doors 3x within 10 seconds. I only hear the normal booting single dings.

Way too much over thinking on this… if you pretty much have a standard day every day. Rarely travel or pull over nighters or have physical problems with your dexterity than rechargeable are for you. If your travel for business may occasionally spend nights away or on the road or just dont want extra stuff a pack of batteries is the way to go.

1 Like

I agree but would add if you do a lot of streaming and am concerned about rechargeable runtime per charge. That was one reason I avoided Phonak.

From support.apple.com:

  • Turning on Airplane Mode, turns off all radios except for Bluetooth. If you turn off Bluetooth while you’re in Airplane Mode, your device will remember that and will turn off Bluetooth the next time that you turn on AirPlane Mode.

That is what Apple devices do but they don’t make hearing aids.

Ah, thank you. I figured that Jim meant he put the Phone in Airplane mode. I don’t see where my Omnia’s have a flight mode. Not important, I have only flown once since COVID and that was private.

1 Like

That was my point repeatedly, too. Especially if going into Flight Mode degrades speech understanding. To save battery, if low on juice, it might be better just to turn the hearing aids completely off until needed again unless frequent rebooting takes more juice that just leaving HA’s in flight mode.

Just as another control, I found it quite easy to put my old rechargeable Quattros in flight mode by holding down the power buttons > 9 seconds. The Quattros then disappear from the Smart 3D status display, don’t show up in the 3D Find My Hearing Aids applet, and aren’t detectable by a BT scanner, unlike my disposable battery-powered Omnia 962s. So, there’s something wrong with my replaceable battery-powered Omnia 962s and flight mode.

The Quattro manual has 14 references to “flight,” whereas the Omnia manual has only six. The Quattro manual mentions the feature has to be turned on by one’s HCP, whereas the Omnia manual neglects to mention that. Neither manual suggests going into flight mode to preserve battery power for critical needs later in the day (having a portable battery charger or using disposable batteries is clearly a better option, as you won’t have to sacrifice the best HA function in flight mode). The Quattro manual also mentions that flight mode may be useful in areas where RF transmission is prohibited, but the Omnia manual doesn’t (there used to be zones where you were told to turn off your car radio because explosives were in use - that might have been vacuum tube days! Turning off wireless transmission at a secure installation might be another consideration).

Thanks for mentioning this consideration, Bruce. Given that thought, it’s amazing that ReSound doesn’t have flight mode turned on by default. If you’re used to leaving your HA function on in flight in the U.S. but then travel to a far-off locale and the flight attendant tells you that you have to put your HA’s in flight mode, you’d be screwed by having flight mode turned off by default in the U.S.A. version of Smart Fit and no HCP available to change the fitting setting. There’d be a lot of frustration before you just decided to turn your HA’s off as required by the flight attendant.

So now, if I ever find myself on a flight taking off in East Globalquerque and am told to put my Omnia 962s in flight mode by the attendant, I know from the start I’ll just have to turn them off! :roll_eyes: