Hey all,
Just came into Phonak Naida M SP. I am currently suffering from mild hearing loss. The Naida has too much power for my hearing loss. Are there any ideas how to reduce the output. Need around 30- 40 db.
Thanks
Hey all,
Just came into Phonak Naida M SP. I am currently suffering from mild hearing loss. The Naida has too much power for my hearing loss. Are there any ideas how to reduce the output. Need around 30- 40 db.
Thanks
To put your audiogram into them?
Yes
My Audiogramm is
L 40 35 30 30 25 30 35 35 35 35 35
R 35 35 30 30 30 25 30 30 35 35 35
You need to program the hearing aids by putting your audiogram into it.
Request the Phonak Target Software and buy a Noahlink Wireless.
The Naida M-SP has way too much power for your mild hearing loss.
Why did you get this Super Power hearing aid?
Directly below you see the Naida M SP Fitting Range diagram
As you see from 1000Hz on upwards the fitting range exceed 50 dB ! And it seems to be not possible to lower the mid and higher frequencys under 50dB.
-----------------------------
For your mild loss an Audeo RIC with S-receiver would fit way better.
Audeo S-Receiver fitting Range
You could also choose Naida Lumity L-PR / Naida Paradise P-PR
This is the fitting range of both Naida Lumity L-PR and Paradise P-PR