I am trying to save my own copy of the hearing aid fittings (for Phonak Naida P70 UP). When the audiologist selects the print option on the Target 8.1 software to generate a PDF software, the numbers on the PDF document are wildly off from the numbers showing up on the screen within the software.
We have tried this out with all the setting iterations, binaural, each ear, numerical, graphical - nothing generates an accurate PDF fitting report!
Never heard of this happening, can’t see how it could happen (truth be told I’ve never actually checked this!) just out of curiousity are your HAs connected at the time of printing, or just from a session that is open?
My guess is you are looking at one measurement standard ie:real ear or 2cc or insertion gain and the software is printing a different standard…see options in pic
Hi @cleveto, thanks so much for your suggestion above. Resolved the issue much to my joy…I now have all my fitting reports saved accurately!
Writing out the resolution below in case this thread is helpful for anyone else in the future:
Check the measurement standard on your setting as per @cleveto’s screenshot
Exit your session and navigate to the home page of the Target software app. Go to settings > reports> there is a drop down list for what the reports are generated in. Select the measurement standard you need. Note that this changes report generation on the software for all sessions.
Go back to your saved session and generate the report