Oticon Opn: Adjustments to boost male deep voices

I put your loss in the Rexton software. It recommended micro vents (1 mm) in the molds. That may be what you have now, but if not, there may be an opportunity in reducing your vent size.

A deep male voice is likely in the 80-160 Hz range. I check the prescription formulas in the software and the NAL-NL2 seems to give the most boost in the very low frequency range. See below. You could ask your Audi if NAL-NL2 might be better than what you have now for bass boost. I also see that the software is recommending frequency compression, but that is unlikely to help with a male voice.

Last you could just ask for a boost manually in the 100 Hz region.

Thank you Sierra,
One more question please:
When my Audi ordered and setup custom mold for me I saw she putted 14mm on Gennie2. Is it to big?

Could it have 1.4 mm that sounds closer to correct.

I think somewhere around 4 mm is getting pretty big. It could be the length, but the standard for length is 16 mm. Here is a graph showing the impact that vent size has on low frequency loss.

image

Some molds have adjustable vent sizes. You could ask if yours could be reduced to 1 mm if it is currently larger. It would help a few dB in the very lowest frequencies where you need it.

The other issue is how well your mold fits in your ear. If there is leakage around the mold then you lose low frequency too.

My molds are hard clear shell. I can see the receiver inside the mold. These molds fit really well in my ear. There is one little hole (the vent? may be?) on each mold. While waiting for next appointment.
Can I get something to plug that hole? (may be the tap?) Then go to Gennie to change the vent size to 1 mm? is that ok to experiment? I guess it does not hurt the HA right?
please advice

thank you again all

Some vent holes are made to be adjusted – Select a Vent or SAV. I think at least one of mine are that way. They came with a package with a bunch of tapered little plastic plugs/vents. I would not have the nerve to try putting them in my molds. Looks like too delicate a job for my fat and aged fingers. It seems like the easiest way to do a quick trial would be to put a piece of tape over the vent and try it like that to see if it improves things. It it does, then I would get it changed at the next appointment by the pro. My fear would be getting something stuck in the vent hole and not being able to get it out again. That said some mono fishing line might do the trick…

Thank you Sierran and others for your supported. I will give it a try to see how this works. I will let you all know

Did you tried wearing aid slightly loose than inserting very tightly? Does it make difference?

Well, the hard shell (or custom mold) can only get in the ear complete or it will fall out. I can not wear half in or half out because it either falling of my ear or it will whistle really really bad! so answer to your question, No I can not wear loosely

If closing up the vent hole in your custom mold still doesn’t help, and assuming that your mold is already well fitting and is not loose and leaking, I think Sierra is probably on to something when he (or she?) suggested to try different fitting rationales as the next step after trying to close up the venting. It’s something your audi can set up easily for you. There are 4 programs, so with VAC+ in the default program, you can use the other 3 programs to try NAL-NL2, NAL-NL1, DSL v5.0 Adult in parallel. Heck, if none of those 3 seem to help, you can also try DsL v5.0 Pediatric next as well.

As for using Speech Rescue to apply frequency lowering to help with your hearing, although it’s most likely not going to help much with low deep male voices, it looks like it’d still be helpful overall because even low deep male voices still employ the “s” and “sh” sounds and such in the English language, so I would suggest having your audi turn it on as well, and set the destination target to the leftmost Configuration 2.4, which has the destination range of 1.5-2.4 KHz.

I am a he! Sierra is actually my dog’s name, and that Sierra is a she!

1 Like

All,
You are “HEROES!”.
Since I cannot have appointment with Audi right away to provide the suggestion from all of you, so what I did were:
1/ cut the rubber band and stuck into the ear mold vents
2/ Went to Genie2 and under the selection tabs, then I changed the vent size to 1mm. (cvKemp was right. my previous setting was 1.4mm!)
3/ change P1 to VAC+, P2 to NAL-NL2, P3 to DSL v5.0 Adult, and P4 is speed in noise. Then re prescribed for all programs
4/ Went to P1 under fine tuning tab, I select all 25hz to 2K columns and increase two clicks
5/ I wanted to change Soft Sound Perception up to increase sharpness but It did put to max already. I cannot change anymore.
6/ Went to open sound navigator changed all P(s) to “Open automatic” selection. and increase max help for speed in complex senario

Well the outcome was: “I can hear extremely loud!” on P1. Basically, I can hear the AC running, refrigerator was humming. fan was running… and other sound from mechanic that I never heard before or barely hear!
I went to work and it seemed I can hear now certain men voices (still not clear and crypts but hey we do make progress right! …and I can hear people talk on the phone. Before I lost certain people who talked to me on the phone.

I am starting to see the light now! but I have few more the questions please advice:
1/ I seemed to loose the sharpness of the voice that I used to have in letter s, sh … (high frequency words) --> should I put back 1.4mm with the vent close?

2/ why P1 was extremely loud but other P (p2, p3) are still in normal range of volume and seemed P2 and P3 were not affected by all setting I had above?

3/ after all of the changes from the above, for P1, the red gain bar went down to 2.5db. and the blue gain bar is still at 7db. How do I bring up the red bar(my right side)?

another note: the sound I am hearing now is like I knock on the plastic box/can. the sound is dull and dry. Before doing this, I heard the sound is sharp like the the sound of the metal can or cookware. Please give me some advice for how to balance those sound

Thank you all again for the helps

Prompted my to change my avatar. No that is not me either. That’s Sierra my dog!

1 Like

You have to finetune also P2 and P3, like you did at P1.
This is the reason that there is hardly difference at these programs.
Fine tuning can be done individually in all programs.

Thank you evert!
Well, the P2 and P3 in my HA now are NAL-NL2, and DSL v5.0 Adult and these are not having the tab “Soft Sound Perception”; therefore, I cannot adjust the sharpness/dull voices .
I think what I can do with these programs are just highlighted all the columns (from 125hz to 8K) and increasing the volumes. I did that but I don’t see much of the different when I increased volume of prescribed graph to above the target dashed line graph… I think when I changed the vent’s size from 1.4 mm to 1 mm, it would affect all programs but I could be wrong. Let me see this evening if I P2 and P3 are having same vent’s size as P1.

thanks again

Reducing the vent size in the software if anything will reduce your bass, not increase it. If you want to trick the software into boosting the bass you would have to increase the vent size in the software but not in your ears.

Well,
I thought earlier I understood that if I need to hear men’s voice (bass) then I need to bring the vent’s size from 1.4 to 1 mm and indeed I did hear some men talk now … but the sound is dulled and dried like the hitting the plastic container… Do mean to to increase vent’s size will increase treble (high pitch voice)?

Aslo, do you know what is this little clock display next to 0.8mm. Is it mean that base on my audiogram, it recommended me to have 0.8mm?

Thank you

I did check all Programs have same vent size. I just set it to 1.4mm(it was 1.4mm before and I brought it to 1mm) with the vents are still plugged up. I will test HA out tomorrow at work to see if it is making any improvement
Thank you all again

I am only familiar with the Rexton software, but I suspect that symbol is the recommended vent size. Rexton recommended 1 mm which is essentially the same. Keep in mind there are three issues at play here. One is the target gain curve which comes from the prescription like NAL-NL2. The second is the vent size. Each vent size has a gain loss associated with it, and in the case of the open dome, a small gain in the midrange. These losses and gains are shown in the venting graph I posted earlier. The open dome has a 30 dB loss at 250 Hz, while the 1 mm vent has only a 5 dB loss. If you just switch an open dome to a totally closed or 1 mm vent, you gain 25 dB at 250 Hz. That is only if you don’t tell the fitting software about it. If you tell the fitting software what you have done, it will try to correct for the vent size change. If you change both the actual vent and tell the software about it, the only real way you gain is if the HA or receiver was unable to deliver all the gain needed for an open fitting, it might be able to with a closed fitting. As you can see from the curves changing a 1.4 to a .8 or totally closed is quite a small gain, but it might be enough to hear it, if the software does not wipe it out.

image

One way or another, if your only issue is male voices but not female, I would focus on the very low frequency range of 80 to 160 Hz, and try and get some gain there.

Thank you Sierra,
Let me test out your suggestion and will let you know how that goes

I found this discussion thread very informative. I was also finding it challenging to understand male voices among other things.

The vent size of my custom ear molds is .08 and I requested my audioligist the change default to NAL-NL2 (with information here) and its made a world of difference for me.

I feel there is room to further adjust middle frequency compression (discussed in another thread in this forum) in my Oticon More 1 HAs for music. I’ll experiment that with my audiologist next. The audiology practice mainly deals with Phonak and are not very familiar with Oticon and Genie software.

Thank you all.

1 Like