When I bought my Phonak Savias, I requested the vent in the ear mold be made slightly larger for a special purpose. We thought this would be okay as long as the larger vents did not cause feedback. However, speech discrimination and sound clarity in certain environments seemed to be lacking. Over a period of three years we made many attempts to adjust compression, volume, etc, to improve performance. Increasing volume did not improve speech discrimination. Also, the aids did not work with my cell phone. It was impossible to determine if the problem was with the adjustments, or a problem with the aids. I thought possibly my hearing was deteriorating and nothing could be done. When an additional increase in volume did cause feedback, the audiologist suggested we reduce the vent size to factory specs. Like magic, clarity and speech discrimination returned. I could now for the first time use the aids with my cell phone. The processors had sacrificed sound performance to prevent feedback. Is anyone else experiencing this?
We experimented with vent sizes (both the physical vent size) and the fitting software vent size on my wifeâs Oticon Epoqs. What worked best for us was a large physical vent size in combination with a different (smaller) software vent size.
I forgot what software vent size we ended up with. But it sounded better with a software vent size that was different from the large physical vent size.
Hereâs a link to a post in another thread. Click -> HERE.
My Tuning Audi enlarged the vents in my earmolds and as he said it would, all became richer more natural sounding. So far no feedback problems.
Errr ⌠so the aids worked well when set up as recommended by the software?
But they worked badly when you (incorrectly) increased the vent size for a âspecialâ reason?
By doing this you lost bass volume due to the vents and possibly treble volume due to the feedback manager trying to prevent feedback.
Iâm probably stupid - but Iâm not sure what you are trying to say in your post.
I am not familiar with adjusting the âvent sizeâ with fitting software.
I had the physical vent size increased slightly so I could insert a plastic audio tube from an aviation headset. The point being that the increased vent size caused the aids to perform poorly. Since there was no feedback, I didnât realize initially that the aids were performing poorly and therefore that the increased vent size was causing the problem.
You are true! Too larger vent makes poor hearing. I use 2 Phonak Savia 311 too.
The original vent size on my Savias was 2mm. The problem occurred when we increased the gain. As we increased the gain, volume increased, but clarity decreased. We reduced the vent size to 1mm and the clarity returned.
I would not say this is always true. The vent size should be selected based on your hearing loss. If the vent size is selected appropriately & the hearing aids are programmed to take the vent size (software vent size) into account things should be ok. There is always the danger of letting too much sound escape from a overly large vent, this will result in reduced low-frequency gain as well as increased feedback (which generally reduces available high-frequency gain). Newer products are equipped with more advanced feedback management that should allow for more gain with less feedback and thus larger possible vent size. The clarity issue becomes large when vent size limits gain for soft high-frequency sounds which allow sounds like âsâ, âfâ or âshâ to be audible.
Conversely a vent size which is too small will result in over-amplified low-freqeuncies and what we refer to as the occlusion effect. This results in hearing our own voice in an abnormal way (usually described as a âboomy echoâ or âhearing in a tunnelâ).
I had ear molds with a small vent. I went to a different audi for tuning and he made the vents larger, making sounds more natural and I felt I could hear better. He lowered the compression which helped with the clacking of the keyboard and paper rattling.
I broke part of a mold and today saw my original VA audi who said lets try a (ear cone shaped thing). It increased feedback which he adjusted and is ok for now.
Iâm not sure what vent size does or doesnât do, but I know it does make a difference.
According to your hearing diagram, I believe that your Savia HAs with 2mm vent work very bad in noisy environment even in calm. You might sometimes couldnât hear someone talking when they sit very near you. You should accept some occlusion with 1mm vent.
According to your hearing diagram, you should wear a mold with 3mm vent.
Sometimes we must accept some occlusion. Because, if you want feel total free of occlusion, you wonât hear well with HAs.
Maybe Iâm stupid. I have ear molds, what is a vent size?
Thank you Zebras. I found it.
There is this excellent review article cited by MDB in 2017 on the effect of venting. Towards the end of the thread I more specifically include a link to one of the papers on which the review is based, a 2006 Widex technical paper that has a graph of a group of mildly sloping high-frequency hearing loss subjects that were tested with CASPA3 words in quiet at a 30 dBHL level. Word recognition scores decreased 30% with increasing vent size.
NOTE: The original link that I provided to the WidexPro website seems to be broken (Widex locked the door except for registered providers only?). I have added another post with a link to a publicly accessible version of the 2006 paper to the thread cited below.