Does Noahlink Wireless doesn`t work with USB C port?

Hello
Noahlink doesn’t work for me with the usb c port with adapter.

Is there any solution?

I suppose you can get a connector to connect Noahlink directly without an adapter. Would that work?

Yeah that’s the way, use the correct adaptor, so what PC doesn’t have a USB port nowadays?
What’s that cable in the picture as well?

Cable looks cockeyed to me, not seated in the adapter. You probably messed with that already, so that’s not the problem.

My guess is the adapter is faulty. But a cable with USB C on one end and micro usb on the other. Don’t scrimp, get a reputable brand.

I have three of these cables, they all work as they should.

https://www.amazon.com/Cable-Matters-Micro-Braided-Jacket/dp/B00UUBRX0Y/ref=sr_1_3?crid=16YVAUGVF7WW6&keywords=usb%2Bc%2Bmicro%2Bcable&qid=1698101933&sprefix=usb%2Bc%2Bmicro%2Bcable%2Caps%2C109&sr=8-3&th=1

You can replace the Noahlink wireless cable with a Type C to Type B cable.

Please note that there are two types of Type B interfaces, please purchase the same interface as on the original cable.

1 Like

That must be the issue, good to know.

I tried 2 adapters. .

It is a Macbook (M2, March Model) with only usb c ports. I am using Windows with Parallels.
Previously I used my macbook (2015 model) with USB ports, Windows with Parallels and Noahlink Wireless was working fine during years.
I already tried 2 usb C adaptors.
I also upgraded the firmware of my Noahlink Wireless but is not being recognized.

Ah ah, okay so Noahlink wireless won’t work with any ARM based PC’s

https://www.himsa.com/support/noahlink-wireless-knowledgebase/preparing-for-noahlink-wireless/arm-based-windows-not-supported-for-noahlink-wireless-or-noah/

2 Likes

Good job.

Not a cable problem at all. LOL.

This cost a little more than a cable, but it should do the job. You’ll still need those two adapters.

https://www.amazon.com/Dell-Latitude-i5-6300U-Certified-Refurbished/dp/B07GCRHS5V/ref=sr_1_4?crid=YQGBS9143DLV&keywords=laptop%2Bwindows&qid=1698158234&sprefix=laptop%2Bwindows%2Caps%2C212&sr=8-4&th=1

It’s a computer problem!!