I’m wondering if you were able to follow those bread crumbs from Sonic?
-
EXPRESSfit 6.4 is used to program products using the Sonic Sound Processing Technology Platform (e.g. Velocity).
-
Products that use EXPRESSfit 6.4 should be programmed with the battery out of the hearing aids.
You will need power to connect/program the Sonic Velocity hearing aids. Actually you need power to connect/program any hearing aid. With the batteries out, you will not be getting power from the battery. Thus, you will need to get power from the cables/flex-whatevers.
Now I don’t know for sure, but I don’t think you can get power with a three-conductor connection to hearing aids. I think you will need a four-conductor connection.
The three-conductor connections that I have used in the past were programmed with the battery in.