@Neville @Um_bongo

Neville I can’t tell you how much I appreciate that information. It explains why I remain frustrated in finding an adequate solution for SIN, especially when others seem to have gotten there.

That audio gram led my audiologist to insist that I have an MRI to verify the absence of acoustic neuroma before she would fit me, especially after hearing that my sister had one. Hers was quite large and it was removed with gamma knife. Unfortunately it cost her about 70% of her hearing on that side. I was clear. My damage, as I noted above was from a blast on my left side. And I suppose you can tell by the speech recognition score that it involves more than volume. There is a significant amount of distortion involved. It’s tolerable at normal conversational levels but as the volume increases so does the distortion. Descriptively, it makes folks sound garbled, as if they were talking from a mouth half filled with water. With the Widex, to achieve some help in complex environments, I would put it Impact Mode, raise the volume a couple of clicks on the right but then had to reduce it that amount or more on the left or the distortion would kick in and become distracting. I got away with that for a couple of years but it doesn’t work anymore. Now I just pick up words here and there. Sometimes I can put things together but more and more lately I’m just watching lips move. The More 1, in spite of many of the stellar reviews I’ve read is no significant improvement, being only marginally better. Complex environments are just a hopelessly chaotic mess. I’m really hoping that the new Roger device provides some significant help.

Any further suggestions?

Um_Bongo how about you? By the way you were right about the Widex. Great sound but the feedback manager sucks. I finally had to go with a closed dome on the left side. I got used to the occlusion.

Thanks in advance

2 Likes

“The most erroneous assumption is to the effect that the aim of public education is to fill the young of the species with knowledge and awaken their intelligence, and so make them fit to discharge the duties of citizenship in an enlightened and independent manner. Nothing could be further from the truth. The aim of public education is not to spread enlightenment at all; it is simply to reduce as many individuals as possible to the same safe level, to breed and train a standardized citizenry, to put down dissent and originality. That is its aim in the United States, whatever the pretensions of politicians, pedagogues and other such mountebanks, and that is its aim everywhere else.”

― H.L. Mencken

2 Likes

Oh I definitely agree with you

1 Like

Serious question: have you tried many RIC aids with mismatched receivers?

More 1 with 60/85?

Resound Ones (the M&RIE) might work by simulating the Pinna effect, even on the worst ear……(again feedback risk)

Starkey Evolv?

1 Like

I’m also trying to figure out which aids will work better for speech in noise, particularly the ability to distinguish between several simultaneous louder voices. Similar to Mark, I have pretty drastic differences in my hearing losses between my right and left, with the left sudden loss occurring just earlier this year.

I have tried both the More as well as the Paradise so far, and personally have found the Mores to be far better in helping me distinguish between multiple noises. If you’re in a scenario that is noisy but there’s really only one speaker in that noise, perhaps the Paradise would work better, as the background noise attenuation is fairly noticeable compared with the More, but in a setting like a restaurant where there is likely to be nearby voices that are loud, I found the More to be more helpful in understanding the person that I actually wanted to understand. The Paradise just created a wall of noise (voices) that blended and was inseparable. I did not try any Rogers accessories with it, perhaps that would help.

I am currently demoing the Pure Signia AXs, and haven’t quite subjected them to a highly complex noise environment as I have with the other aids yet (I intend to), but as far as their performance in noise is concerned, to me, they sit somewhere between the More and Paradise. With the More, I felt that the soundscape was more “natural” to me (there’s a whole thread with a bit of debate as to what that means to each individual), but the overall environment sounded closer to what my hearing was like prior to my drastic change in my left ear earlier this year; there is less attenuation of background noise, yet speech is relatively defined or focused through that noise. Though there were still times I felt like the overall sound environment was a little more heightened than what would naturally occur, thus sometimes distracting (though it could be attributed to the fact that I have in fact have had hearing loss my whole life without aids until now).

The Paradise creates more of a separation, thus making it sound more unnatural to me, but if someone’s sole focus is to be able to hear actual speech better, than I think it certainly accomplishes that. My issues with these aids, as mentioned, is when there are several loud voices all in the vicinity. While background noise doesn’t get in the way, it’s distinguishing between those voices that becomes nearly impossible for me. There were several instances in which I actually heard better after taking out my HAs. Seems quite counterintuitive. I suppose it just really shows how difficult it was for HA tech to truly mimic the auditory process of the ear and brain.

I’ve only had the Signias for just over a week, but they do seem to be middle ground between the More and Paradise, probably because of their dual-processing approach between voice and noise. Voices do seem to be heightened and brought to the foreground, yet background noise is still distinguishable and audible. The sound seems layered, so still not quite as “natural” to me, but not unpleasant either. Because I have the Paradise for trials and was given the AX as a demo, I currently have both of them, and have been able to do a couple A/B comparisons. The layered effect vs. the strong beamforming/attenuation effect is definitely noticeable in certain situations (at a busy dog park, for instance), but in other situations, the difference between the two are less prevalent. I set up my laptop and played a random noisy YouTube clip, and tried to watch a recorded show on my tablet and tested which aid might help me hear the show the best. I’d say the AX slightly edged out the Paradise, but to be honest, without the immediate A/B comparison, there wouldn’t be much of a way to definitively tell the difference.

Given that our personal experiences and preferences will differ, I would recommend trying the AX too! I’ve seen very little actual user reviews of the Evolv so far, would love to find some.

I have not. I was hoping the Roger streaming directly to the aid would work.

To a huge degree, the ability to segregate speech streams in a noisy environment relies upon one’s ability to localize them within the space. Blast damage appears to impact structures beyond the cochlea and into the auditory system more significantly than a chronic lower level noise exposure. Your speech clarity on that side is low and binaural timing and level differences are muddled. One can localize sound monaurally with spectral cues, but the weak spectral cues occur up above 2 kHz and strong ones up above 4 kHz and you’re losing ground there in the good ear. I suppose all that is just to validate your experience that your hearing loss is different than someone with a similar level of more traditional presbycusis.

Modern premium hearing aids claim to be able to provide an SNR boost of 6-9 dB, but keep in mind that this is in very specific controlled laboratory environments. A real-world value might be more like 3-5 dB. It is not uncommon for people with hearing loss to require a signal to noise ratio above 8dB, and it is not uncommon for a noisy restaurant to have negative SNRs. So at that point, yeah, it doesn’t matter how fancy your hearing aids are. Clipping a mic to your conversation partner’s collar is always going to give you the best SNR, which works for one-on-one but not so well for groups. The roger on will give you a better boost in a group but still won’t solve the worst noise environments, but at least you have control of who it is close to. Look for environmental solutions as well–do you have any control over when and where you are meeting people? Often locations for social gatherings can be a bit arbitrary and groups might be open to suggestions of a smaller restaurant or a less busy time. Can the family member with the more traditional carpetted dining room host more often than the one with the open concept hardwood great room? Sometimes this sort of thing can be really low-hanging fruit compared to eking out another quarter dB SNR with hearing aids.

6 Likes

@Mark_Chambers: What kind of civilian range would permit firing an 8,000+ foot pound round?

But, I think it’s incorrect to class the muzzle blast of even the mighty 460 Weatherby with the kind of overpressure blast waves that artillerymen and IED victims have been exposed to. It’s the overpressure waves that cause the deep structure damage to which the audiologists are referring, I believe. [NOTA: I have little to no expertise in the physics of small caliber muzzle brakes, admittedly, so my conclusion may be erroneous regarding the damage that the infrasonic component of the blast can cause.]

I could be wrong - one of the Providers will correct me if I am (I hope).

@SpudGunner

“I have little to no expertise in the physics of small caliber muzzle brakes, admittedly, so my conclusion may be erroneous regarding the damage that the infrasonic component of the blast can cause.”

A brake is a vented device that installs on the end of the barrel and takes the blast pressure that normally vents out the front and redirects it outward and rearward, more or less perpendicular to the barrel. Unfortunately that includes the report. Many private clubs will not allow braked rifles under covered benches. This was many years ago at an unregulated public facility. Here is a short quote from a Field and Stream article.

"• They Will Deafen You (If You’re Not Careful)

The biggest problem is noise. A muzzle brake will deafen you if you are not very careful. With the muzzle blast coming toward you instead of traveling away from you, the report goes from unpleasant and ear-damaging to unbearable and ear-killing. You can’t shoot a muzzle-braked rifle without earmuffs, period, even when you’re hunting."

Shooters are normally extremely considerate and polite. Unfortunately it only takes one asshole one time to cause damage.

1 Like

Yes, I know what a muzzle brake is, Mark.

2 Likes

Couldn’t tell you. What I can tell you is that it was physically painful. It hurt like hell into the next day and the ringing was there for at least two.

When questioned by the audiologist when I was tested and fitted for my first set it was the only thing I could think of that would result in the difference between the right and left. Hearing issues run in my family. My father and every one of his 7 siblings was hard of hearing. They were also bull headed and refused to wear aids.

I hope you don’t think I’m implying that the muzzle blast didn’t deafen your left ear. It would be enough. In fact, I’m surprised that it didn’t do big time damage to your right ear as well.

No such thought entered my mind. As far as the blast it was directly to the left with nothing between us but some space and I felt the pressure and the heat. But the noise was deafening. Literally I guess, HA! Who knows for sure. But It’s the only thing I can think of that was unique to the left side.

2 Likes

Sorry to bring this old topic back. I have a question about the More’s Speech in Noise program.

DSL is a no brainer for my music program. I am very happy with it and my acoustic and classical guitars sound great. However, I have been going back and forth between VAC+ and DSL as my main program (slot #1). Problem is that the Speech in Noise program also changes every time I change between VAC+ and DSL.

Let’s say I am comfortable and happy with either a VAC+ or DSL main program. Which fitting formula gives me the best Speech in Noise program? Since VAC+ is Oticon proprietary, would it yield a better Speech in Noise program compared to DSL? How much “automation” goes on in those “stock” programs Genie offers us? Can I get the same level of this “automation” using DSL? Genie 2 doesn’t allow us to have a Speech in Noise program based on VAC+ and a Speech in Noise program based on DSL at the same time, right? That makes it hard to do an A/B test. Would @Neville, @Um_bongo or any forum member have any thoughts on this?

You can have both, you just have to hack it. The More “Speech in Noise” program is just whatever is in P1 with the noise reduction features cranked up. You can copy your DSL (or VAC, whichever is not in P1) program and just crank up the noise reduction features to the same settings that the “speech in noise” program uses.

That said, the VAC+ has a floating linear window that is not maintained in DSL which may or may not make a difference to you (though, if you don’t notice much difference between a VAC+ and DSL main program, then probably not). Additionally, DSL has separate noise targets that were introduced a few years ago. I know that they are available on the audioscan real-ear machines, but I’m not sure about others. Your clinician may or may not be aware of them. They are not available in the software, but they are just gain changes so you can start with a DSL base and verify to noise targets. You can equally start with a VAC base and verify to DSL adult noise targets if available, and maintain that floating linear speech window. Offhand, the DSL noise targets are probably somewhat in between the VAC+ targets and the DSL calm targets.

Thank you, @Neville. Lots to unpack here, gonna do my homework and learn a bit more about what you’ve said :slight_smile:. Two things I noticed. First, the “Speech in Noise” program is louder than P1. Second, P1 recommends sound enhancer to “Balanced”, while Speech in Noise “Detail”. That puzzles me, I thought this would be the other way around.

@e1405 → I tried looking at the gain curve for my type of hearing loss for the More 1 on Genie 2, toggling between the P1 default and the P2 Speech in Noise program that’s a built-in program from Oticon, so I can compare the 2 gain curves visually on-the-fly to discern their difference more easily. And indeed the gain curve of the Speech in Noise program is slightly elevated between 500 Hz to 2 KHz for my hearing loss. What this is implying is that the Oticon built-in Speech in Noise program is not just exactly the same P1 gain curve but with the noise reduction parameters and such adjusted to make it Speech in Noise, but Oticon also seem to have made some slight adjustment on the gain curve as well. So while you can set all the parameters in the MoreSound Intelligence menu (like the noise reduction and the other stuff) in P1 to match exactly like what they are set in the Speech in Noise program, the fact that the 2 gain curves are slightly different implies that you won’t get the exact same result, but probably close enough.

But @Neville is right that if you want to have an equivalent of Speech in Noise based on DSL instead of the default Speech in Noise program which is based on VAC+, just copy all of the parameter values in MoreSound Intelligence, and whatever you have set in Automatics and Speech Rescue (if any) that appears on the Oticon Speech in Noise program over to your DSL program, and you should have the equivalent of Speech in Noise for DSL.

The screenshot below is the help section for the Sound Enhancer in the MoreSound Intelligence menu. If you read this, it seems to make sense that in the Speech in Noise program, you’d want the Detail setting for the Sound Enhancer so that you can discern speech over other sounds more easily. But in P1 which is for most other environments, you don’t need that kind of level of details that you may need in Difficult, so the Balanced value probably is a good compromise between Comfort (which would lose the level of details but make things sound more smooth and comfortable) and Detail (which would make things clearer but maybe at the expense of louder, sharper, and more tiresome sounds to listen to).

Sorry, when I looked I didn’t see that. But it may be hearing loss specific and just not apparent with the curve I had in there.

Thanks, @Volusiano. Now it makes sense to me. It seems that the Speech in Noise program boosts the important frequencies for speech while tuning down everything else.

From what @Neville explained about VAC+ and DSL, and the discussion you are part of in the REM vs In-Situ thread, proprietary rationales optimize sound processing in ways the manufacturer deems appropriate. It has been said on this forum that the goal of the optimization is to streamline first fit. However, I wonder if the analogy I once used (iOS/iPhone vs Android/countless different hardware) applies to HA and proprietary formulae as well. I mean, there is a lot more going on other than what we see and are able to control in Genie 2. Oticon calls this “automatics”. Hence my question: since the Speech in Noise is a built-in program, would it work best with Oticon’s VAC+ template? If I have VAC+ on P1, would I have the same “automatics” going on in a DSL program with the exact same parameters of a “cloned” Speech in Noise program? If I have DSL on P1, would the Speech in Noise program be constrained anyhow because that is not Oticon’s formula? Those are questions I could answer empirically, I suppose. Perhaps this is even a fool’s errand (maybe not, since I do hear 1dB sound increase)… Just asking them out of curiosity at this point…

I’d have to double check whether this is still a thing Oticon is implementing with the More, but historically my understanding is that the answer would be yes with DSL but no with NL2. This is only because of that floating linear speech window that Oticon likes is coming into conflict with the slightly more prescriptive standards that have to be met by the manufacturers in order for them to include DSL among their options. Because that linear window did not meet the compression rules in DSL, they had to turn it off. Note that this floating window is a nice idea with no independent evidence for benefit and which did not seem to make a slam dunk difference for Oticon. I’d have to double check whether it is still active in the More processing or whether they dropped it with the AI. If they dropped it, then it shouldn’t make a difference–I can no longer see it listed in the features, but it was there in the Opn.

Note that the independent standards only define the provision of gain. They do not make rules about attack/release times or automatic features, so typically (lately) those are all active and the same across different rationales except in specific instances of which Oticon is (was) one (for DSL specifically, not NL2).

1 Like