Looking for sifit database for connexx6.4.5/7.2.2

Thank you very much Ricked

Looking for connexx 6.5 database for REXTON RX38 TARGA PRO 113/47 please help e

Hi. Is this link still working! I couldn’t get it. Any help would be useful.

Which links are you after?

He’s looking for this;

Hi, got the link, installed, could see the aid in database as well. However I could not detect the aid. I used flex, cS44 cable and HiPro 2.

Are you using a battery in the hearing aid for power?

Oh, the cable guide shows that RICs need the battery in. I’m unsure if the Nitro needs the battery in?

What about the Black Alignment Dot on the FlexStrip? Is the Black Dot facing downward toward the battery door like in this picture?

ETA: Here’s a link to a similar Topic;

it’ll need the battery to power up, and good point on the correct way to align the flexstrip, been caught with that before myself.

Sorry for delayed reply over here… The three strip contacts on the aid seems to be broken, Hence the device was not detected. The problem resolved. Thanks a lot for the help.

1 Like