My wife has the Signia Insio, which is ITC type. In reality, only half of it is in the canal. Other makers would call it Half Shell.
We are buying from Specsavers at very good price compared to ‘independent’ providers. I don’t think the audiologist would be prepared to do some lenghty experiments.
I have read somewhere that most audiologists would not consider fitting for music to be part of their role, and would not know what to do.

May I suggest your wife buys a pair of Apple iPods Pro? These have the ability to add a compensation curve based on an audiogram. I have Phonak, but if I want to listen seriously to music I don the iPods. At around $200+ they are far cheaper than a HA! Not as convenient swapping though, I’ll admit. Note that a new vcersion of the iPod Pros is apparently imminent. Apple seem to be seriously investing in the market for people with hearing loss.

This sucks. It could be true, but if it is it sucks. Music is a big part of hearing. I know a great musically inclined clinician in New Zealand if you want to go on a trip.

Is your wife’s hearing loss the loss that is in your avatar?

There are some very well understood rules about listening to music with hearing aids. You can easily find them if you Google, but the ones I remember off-hand are:

Use a specific music program on your hearing aid.

In that program:

  1. you need feedback suppression to either be off or set to a specific music setting or at least set to the minimum that prevents feedback.

2.Turn off any frequency compression, frequency shifting or “sound recover” type of technology.

3.Turn up the MPO settings.

4.Keep the level compression as low as you can, but obviously avoid clipping.

5.Turn off noise suppression, noise limiting, wind noise suppression and other such things.

Listening with just one ear, are there noises that sound particularly prominent? If so you could ask for those frequencies to be reduced a bit. Do the same with the other ear.

I find that you can massively improve things by doing this. There is no reason why the settings for music should be the same as the ones for general conversation.

Also lastly some hearing aids work much better for music than others. I like my current Resound Ones, but no doubt others here can tell you others that work well for them.

It’s true that hearing aids can never give you back the quality of sound you have before you needed hearing aids. But you certainly can get to a place where you can still enjoy music. It needs some work though.

4 Likes

I work for Specsavers in Australia and I would (or have a go at it at least). Just ask the clinician if its ok.At a guess I think it would only take 15 minutes or so to tweak a dedicated music program - Just let them know in advance so they can allow a bit more time. Lower price shouldnt equal less service. Good Luck

2 Likes

So is it true that you guys lock your hearing aids, if so why? Or
If it’s just simply a custom version of target software, are you willing to share it so that perhaps people can DIY.

Jfranek43 - you’ve gotten a lot of advice and commentary here. There is a simple solution, touched upon by David.Hendon above, that has worked VERY well for me with my Oticon Opn 1 aids.

Setup a program with your audi with just your frequency adjustments and no other processing. This should give at least a good baseline to tweak further, or in my case left me very satisfied. I use this “Music” program happily with both speakers and headphones - the only caveat for me is the headphone pads must have large enough holes to fit my largish ears.

I found this information after extensive searching, on this site if memory serves.

Yes, Neville. I am posting here on my wife’s behalf, and the chart in my profile is her hearing test.

That is what I intend to do - ask the audiologist to wipe any other settings out.

Interesting thing - today, my wife is using the Recorded Music program throughout the day. She does not notice any difference from the Universal program, and neither am I or our daughter when we communicate with her.
I don’t know what to think of this. Either the music program is a copy of the universal program, or my wife does not need speech recognition algorithms to understand speech in quiet environment.

I am sure the audiologist will try to do it. We are going there after lunch - she said it won’t be as busy as mornings.
At the last session, we spent most time trying to get the Signia Assistant installed on my wife’s mobile. The audiologist could not manage to do it. .
She rang Signia support, and the guy told her the Galaxy A71 wasn’t compatible, so it won’t work.
Galaxy A71 is not on Signia’s list as a compatible phone for Android direct BT streaming.
That list is out of date.
A71 is compatible, it supports the ASHA protocol. Direct streaming works on it.
When we got home I re-read a Signia paper about how to install the Assistant, and managed to do it. The trick is that the hearing aids must be paired with the phone not via the phone’s settings, but from within the app. So first connect with the OR code, and once the FM connection works, go to the Settings of the app, and connect with BT.
The BT connection gives a few more controls, and a mask mode. And the Assistant.

I like your comment that lower price should not mean less service.

1 Like

There is an Aussie product called Nuheara IQ2Max buds, they also have a test, and connects with Android. They also have microphones, so work not only for streaming.
But that is another AU$400 or so and complication with taking one device out and putting another one in. She needs the proper hearing aids for conversation.
My wife (and myself) listen to music with over-the-ears headphones. I use APO equaliser to fix my milder hearing problem when listening from my PC, or an app with EQ when streaming from my Android phone.
Trying to set up an EQ on my wife’s laptop is difficult. She is not technically minded, and does not have the patience to fiddle with it. But I will try if we can’t succeed with an acceptable music program on the Signia HAs.
BTW - Nuheara is also supposed to bring out a new model IQ2Pro soon. I may try it when it comes out.

I think it depends on a lot factors, including the type of loss and type of music.

For me, over-the-ear headphones with noise reduction (limiting external noise) and plenty of volume. I use my in-the-canal HA set which must have a higher noise tolerance, since my BTEs keep cutting out.

But it’s still just not the same. I remember many higher-pitched keyboard sounds that were there in my youth but now are missing from the same tracks.

Best of luck

It is absolutely true that the music quality coming from a laptop or smartphone is like already 2nd or 3rd generation. It’s simply never going to be the same as hearing music live at a concert/symphony for starters. So there should be some “expectation level” setting when one wears aids given the sound wave compression and crappy music streaming from who knows what source.

I also had a pair of Sennheisers that not only worked great with my laptop, but also on phone calls when I was in an awkward transition to an earlier Oticon OPN that had NO phone streamer out for over a year.

The basic nature of how HAs process sound is going to render that inferior to perfect hearing for NOW. Maybe technology will catch up - not sure how if frequencies have to be compressed.

We are not used to listening to music in concert halls or opera houses, so we are not comparing with that.

My wife does not listen to the laptop via its tiny speakers. She uses wired Sennheiser HD599 headphones. Laptop Realtek audio processing is not bad. In any case, she is not very demanding. Mostly watching ballets or opera/operetta clips on Youtube. Andre Rieu too. Unfortunately, when she says music sounds bad, she can’t describe it in detail.

There are many success stories on this forum where musicians managed to configure satisfactory music programs in their HA. We’ll try that, and if that fails, we’ll have to try to figure out a good EQ setting for her.

Ah, that’s a substantial hearing loss for a first time hearing aid user; adaptation is going to take some time. I would strongly consider getting custom tips made. But in terms of being successful, unless something happened that caused her hearing loss all of a sudden, she has to overcome decades of auditory deprivation–like strengthing a leg after decades of wearing a cast. Things may sound sharp and intrusive for a while. That level of loss over 2kHz is also more frequently associated with a loss of frequency acuity which may or may not lead to pitch-smearing that is in the ear rather than the hearing aid, and can impact musical quality. For the music program to sound better in the immediate, you might need to back right off of those highs–she may be able to add them back in over time. If she’s not a serious audiophile, I do still think it should be reasonable to expect that the clinician can get music sounding ‘comfortable’ through the hearing aids. ‘Nice’, even. It helps to have music that she is familiar with in the office while adjustments are being made, and yes it helps if she can think about how to describe her perception.

A lot of the automatic features turn off in a truly quiet environment. Most of the algorithms in hearing aids are attempting to manage background noises and babble. Keep in mind when you do end up going somewhere busy that there are also limits on what a hearing aid can do in noise, particularly when a hearing loss gets to the level of what you have posted. Hearing aids on their own can boost the signal to noise ratio by maybe 4-6 dB. It is not uncommon for individuals with that level of hearing loss to need a signal to noise ratio boost of 9-12 dB to be able to understand target speech in noise, and that sort of boost can only be accomplished by a remote mic system.

1 Like

Thanks, Neville.
Things are not as bad as the test chart indicates. We met in a restaurant with friends before Christmas before she had the HAs and she could converse with people sitting next to her.
At home, it was worse, because a lot of ‘conversation’ is done in passing, not face-to-face, or even talking from another room.
In this regard, the HAs are a big improvement. I don’t think the (expensive) remote microphone will be necessary. We only go to restaurants a few times a year meeting with a few friends.
Meeting with our teenage grandsons is a bigger challenge. Their speech is a pool of mud with American accent. Their father is American. I can understand about half of what he says. He speaks clearly only when he is asking Siri something. The Signia app has a face mask mode - it may help to listen to them. :smiley:
Custom tips? The Insio is all custom. No tips.
Regarding music - as I said before, she has distortion even without HAs.
Her hearing loss worsened from mild to what it is now within - I would say - 6-10 months.
We will try to compensate with an EQ for headphones if the music program is no help.

1 Like

Ah yes you’re right, I was still thinking she was in RICs for some reason.

My hearing loss is similar to your wife’s though a slight bit better at 4KHz and above. Like your wife, music is key to me. After trying several OTC units and two audiologists prescribed RIC units, I have found the Insio AX7 ITC the best for music. RIC units were a disaster for music as their mics(receivers) were on the top of the ear not in the ear. Because of this the pinna effect was lost and music was flat and lost all depth and sound stage…stereo or surround became monophonic. My first pair of Insio’s came in with a poor fit and venting that was too closed. My audiologist returned them for a redo. Signia is apparently not as familiar with fitting as other manufacturers…if fit properly they should be well within your wife’s ears. My audiologist has been very patient as we are on this journey and has been working to optimize my Insio’s for music as well as speech.

1 Like

The patient is referring to a Signia branded product, not the Specsavers Advance. In answer to your question though, no they do not lock their hearing aids - if another clinic requires the software to adjust them, Specsavers will send it out to that clinic. And no I cannot share this software for DIY

Yes I was aware of that,the Connexx software and most others are freely available, this is why my question was directed at you, as you mentioned your with Specsavers, but anyways, thanks for letting us know about the software.

2 Likes