A persons Dynamic Range of hearing is defined as the amount of sound in decibels that they can hear without discomfort or distortion. Normal range is on the order of 100db (SPL) and a person with hearing problems could be much less than half of that.
A vast majority of top line aids on the market use WDRC (Wide Dynamic Range Compression) to handle the decreased dynamic range of those with moderate, severe, and profound losses.
Persons with these sensitivities of loss can only comfortably handle a limited amount of loudness from soft to loud because of damage to their hearing system called Recruitment).
WDRC aids are adjusted so that when the incoming sound reaches a certain loudness, called the Knee Point, the louder sounds are then amplified on a non-linear basis…meaning that softer sounds get more amplification than louder sounds.
It does take the aid a short but finite time, one to ten milliseconds, for the aid to determine if the sound has reached this Knee Point (Attack time).
All aids on the market have some kind of final maximum loudness circuit to limit peaks. But these limiters will cause severe audible distortion if driven too hard. An occasional peak clipped will probably not cause noticeable distortion .