Yep, I get it that for this specific instance of your bricked hearing aids;

  • factory reset using iCubeII with Phonak Target 8.? does not work
  • factory reset using iCubeII with Phonak Target 6.2 works

Also at issue is why were the hearing aids bricked in the first place? You said;

  • I was not even trying to do the firmware as the firmware has not been updated for a long time.
  • I had a new battery in it as I always do.
  • I’m going to stick with 6.2 as l only have the V70 and M70 so it’ll work for me.

So, back to my point I was trying to make. These are Phonak/Unitron version issues related to their going-obsolete proprietary wireless hearing aid programming devices, and I believe other members should be aware of these issues and the workarounds. Not just your workaround. But all of them.

  • Phonak Target 6.2 can fix some bricked hearing aids
  • Phonak Target 6.1 can facilitate the use of obsolete iCube
  • Unitron TrueFit 5.0 can facilitate the use of obsolete iCube

@pvc

6.2 can fix a bricked Aid with the same fault that only works with Noahlink Wireless if you have the Marvel according to the audiologist so again nothing to do with the iCube II, it’s software based.

You’re thinking too much into this. You’re making this thread into something complicated.

1 Like

Software based? So your hearing aids got bricked by Target 8, and the bricking has nothing to do with iCubeII? Really?

Wait a minute? Shouldn’t everyone try putting a stop to this software based bricking of hearing aids by reverting back to Target 6.2? Whoa, wait another minute? I don’t think that Target 6.2 can program Paradise hearing aids. What can we do?

My logic that I just stated is Silly, huh? Your logic is too. Because you are leaving out the hearing aid bricking event, in order to make an equally silly point that it’s software based.

Read again and then you may understand.

How can it be an iCube II issue when the same happens to Marvels and NL Wireless?!

Like I said before, I was talking about the Marvels. Mentioned nothing to do with Paradise.

The audiologist was only helping me with my V70 and M70!

Are you saying that Target 8 (using Noahlink Wireless) can brick your hearing aids? Of, course not! You are ignoring the bricking event to bolster your argument.

Dear Readers; Move on. Nothing to see here. That is, until your hearing aids get bricked by a going-obsolete proprietary hearing aid programming device, namely iCubeII.

No I’m just passing on information from an audiologist of nearly 40 years!!!

He’s more likely to be correct than a user like you! I realise you have a lot of info but come on, 40 years of experience!!

You’re the only one who’s not understanding this situation!!

Honestly your so out of touch on so much regarding nothing you actually know about, you, with the experience of programming 1 or 2 sets of HAs in you life, Oticon at that! you’ve never used anything else, you have no “experience” at all to make any of the clams you do, your a "book reader "you can brick HAs with a Noahlink wireless (and you can brick a Noahlink wireless) and a HiPro for that matter, using any software from any HA manufacturer when you don’t do things right, honestly I just wish the mods here would just “ban you again” god only knows why they allowed you to come back!

1 Like

Please answer the question! It’s a simple question.
How did the hearing aid get bricked?

Specifically, which version of Phonak Target and which wireless hearing aid programming programming device was in use when the hearing aid got bricked?

Let me guess; It was the latest version Target 8.0.3 and Phonak’s propriotary wireless hearing aid programming device (iCubeII), right? And, even though Target 8.0.3 has the ability to to use the “Reset to factory settings” it would not work to UN-BRICK the hearing aid.

The solution from Zebra’s learned Audi (or more likely as a result of the Audi’s convo with Phonak technical support) was to revert to Target 6.2 and use the “Reset to factory settings” tool from this older version to UN-BRICK the hearing aid.

Com’on, you can answer the simple question about how the hearing aid got bricked. No need to resort to name calling and Ban requests. Future events will flush this out anyway. So why not discuss it now?

It’s already been discussed, and the solution, but as usual your not getting it, so now, who really cares/knows for sure what actually happened, but @Zebras got a good outcome to a very stressful situation,end of story.
“Quoted from someone else”
But you always seem to be putting contentious words in people’s mouths, as if you’re looking to entrap someone so you can engage?

A true Jekyll and Hyde character if I’ve ever met one, your right about one thing tho,I shouldn’t have called you names!

1 Like

More obfuscation is not the solution. Hearing aids are getting bricked. Why?

Inquiring minds want to know.

Solution? That’s not a solution. That’s a workaround. Will future Phonak Target versions fix the problem and not brick hearing aids. Or maybe remove iCubeII as a programming device?

@Zebras
@tenkan
@pvc

let’s have coffee together. !!
:coffee:

1 Like

:coffee: LUWAK is good! Let’s enjoy ourselves in Bali!

I’m not drinking that! :wink:
c3

1 Like

Oh wow,no wonder the coffee tasted so discussing when I was last there :nauseated_face:

1 Like

Luwak is an Asian Civet cat. Due to internal enzymes, the taste of fermented coffee is bitter. We even have Elephant coffee in Thailand. Hope you guys come to Asia for holidays ( extra batteries) - Sun, Sand & Sea…HAPPY NEW YEAR!

1 Like

This is exactly what the audiologist of nearly 40 years said.

:+1:

1 Like

Forest for the trees, Detective. Forest for the trees.

1 Like

Yes, you can try using Phonak 7.0.7 (like the airplane). Though, we don’t actually know which combination of (Target fitting software & iCubeII) is best for avoiding the bricking of your hearing aids. We only know which combination (Target 6.2 & iCubeII) was successful in performing a factory reset/A.K.A unbricking the hearing aids.