I wonder if the mini-charger would fit in one of the UV/dryer boxes along with your aids? A small hole with a sealing grommet for the power wire to come in. Problem solved…
I just went through a similar thought process. As it turns out, my HAs (Rexton Adore) come with a charger that has a drying function. Not that this fact was particularly front and center; I stumbled upon it deep into the manual.
I don’t have HA’s with rechargeable batteries, and for various reasons I would not buy them. That said, it seems to me that if the rechargeable batteries stay inside the HA body when they are being recharged there will be enough waste heat generated to warm them up and dry them out, providing there is reasonable ventilation to get rid of the warm humid air generated. In other words if the recharging system is well designed there may not be any need for anything more.
Not sure that the dose and duration of heat is sufficient given that there are few chargers that are specifically designated as drying also. I am aware of only my charger that is so described.
If any charger could also dry, I doubt that manufacturers could resist the opportunity to add drying to their list of “features”.
My dryer only runs for 20 minutes or so. I usually use it while showering.
Not to sound too cynical, but it’s really to the HA manufacturer’s benefit if hearing aids DON’T last as long as possible, so I doubt if drying them out is a priority with them.
This is not true. Maybe true for a mild to moderate loss. This entirely depends on Bluetooth usage and, in my case, the amount of power required by the receiver for one’s hearing loss prescription. My 13-T Marvels only last me 5 days with ultra power receivers.
What’s the humidity like under these conditions? If you look up a Mollier diagram (there’s an Android app called that, too), you will find that at any humidity up to 83% there will be “much evaporation” and from 83% to 97% humidity, there will be normal evaporation at a dry bulb temperature of 32.4 deg C or above. At any humidity below about 68% with temps in the mid 30’s C, you will have “Very Much Evaporation.” So you’re essentially living in a big dry box in central Italy. Drying your HA’s out just by sitting out will ~prevent microbial growth outside your ears. Since you’re not sterilizing the inside of your ears, when you put the HA’s back in, there will still be the same microbial fauna there that there was in your ears yesterday. The most important thing is probably just to use the same degree of cleanliness with your HA’s that you would use preparing your food, e.g., wash your hands before handling, work on a clean, new piece of paper towel or a clean cloth, etc.
Can anyone quote chapter and verse from an OEM HA manual on the need to dry HA’s in a dryer or sterilize them with UV (or by any other means)? My ReSound Quattro manual has no such instructions other than not to expose the HA’s to liquids or humidity over 90% - and it even mentions exposure to direct sunlight (UV) can lead to degradation of the plastic components in the HA’s.
Instructions for Quattro Maintenance (p.34 of RIE961 manual)
- Keep your hearing aids dry and clean.
- Wipe the hearing aids with a soft cloth after use to remove grease or moisture.
- Do not wear your hearing aids when putting on cosmetics, perfume, aftershave, hairspray, suntan lotion etc. These might discolour the hearing aid or get into the hearing aid causing damage.
- Do not immerse your hearing aid in any liquid.
- Keep your hearing aids away from excessive heat and direct sunlight. The heat may deform the shell, damage the electronics and deteriorate the surfaces. (emphasis mine)
- Do not swim, shower or steam bathe while wearing your hearing aids.
(note for list item 5-rechargeables might be more sensitive to the heat exposure component than disposables)
Every morning any wax that was on my molds when I removed them the night before is always very dry-I don’t use a dryer or a sterilizer and I wear totally occlusive molds. Letting the molds dry out overnight makes them easier to clean the next day and I don’t need a dryer to do it. Where I live in Texas, the humidity can often rise above 80% at night outside, especially in the spring and fall, but I’ve never seen the humidity go above 70% inside our house and it’s usually in the 45% to low 60-ish% humidity range. According to the Mollier diagram app, even at 16 deg C, there will be normal evaporation at 65% humidity. We heat and air condition our house very little. I should imagine that anyone who heats and air conditions their house like the typical American has an even drier indoors and even less need for a special dryer since they’re living inside a big dry box called the American house (unless you deliberately rehumidify the air as some do).
@jim_lewis Agree with almost everything you said and certainly the main thrust of it. However you lost me a bit about letting your hearing aids dry overnight preventing bacterial growth. Source? It’s not that I think bacterial growth is a big issue, but I’m skeptical that there’s any evidence that letting hearing aids dry out is enough to stop it. I would think earwax, even “dry” could be adequate growth media for some bacteria. But to your main point, I agree that people worry way to much about drying their hearing aids.
Here’s a Wikipedia article on drying to either inhibit or kill microbes (inhibiting, i.e., “preventing,” in food preparation and “killing” in dealing with sewage sludge): (life depends on water) - some bacteria can enter into a spore state on drying but since that’s a dormant state, one is certainly inhibiting further growth by drying such bacteria)
Edit_Update: The following info about “water activity” in foods vs. microbial growth is interesting. The second table shows how dry, in terms of water activity, something has to be to inhibit various types of microorganisms. Perhaps we could consider fresh ear wax to have the approximate fat/moisture content of peanut butter. But at any rate, the table shows that something does not have to be completely dry to inhibit most bacterial growth. It only needs to have its water activity sufficiently reduced.
http://www.foodsafetysite.com/educators/competencies/general/bacteria/bac5.html
Further_Edit_Update: Don’t know how sound any of the studies are but here’s the Wikipedia take on any possible antimicrobial properties of earwax:
I’ll buy that “drying” reduces bacterial growth, but I’m not convinced that overnight drying is enough to do it. It may be, but I think overstating claims lessens overall credibility (my bias)
As a teenager I spent many days swimming in the ocean surf with goggles and flippers. My ears were wet most of the day. I dried out when I came out of the water or went home at night. The water wasn’t chlorinated and to the best of my recollection, I never got an ear infection from this activity. I did get an ear infection as an even younger kid who stuffed cotton in his ears to avoid hearing his parents yakking downstairs at night, then went swimming, forgetting to remove the cotton, and having in the wet mass in my ears for many days (and nights) straight. So the degree of moisture exposure vs. drying out is very important as well as the bacteria/mold/microbes, etc., you happen to be exposed to.
Edit_Update: to go along with links in my previous post, here’s a quote out of the Wikipedia article on “Swimmer’s Ear:”
" The two factors that are required for external otitis to develop are (1) the presence of germs that can infect the skin and (2) impairments in the integrity of the skin of the ear canal that allow an infection to occur. If the skin is healthy and uninjured, only exposure to a high concentration of pathogens, such as submersion in a pond contaminated by sewage, is likely to set off an episode."’
Was just looking at the Wiki article on antimicrobial effects of earwax and was considering sharing.
I am a newbie with a question following on from a little earlier in the discussion.
My bargain priced Phonak Audeo M 70 R hearing aids were delivered with two compact chargers but did not include the commonly supplied larger charger which has space for the small circular desiccant container.
What to do? Well I recycled several pill bottle silica gel sachets (after reactivating them) and I insert one of them whilst charging the HAs. I was not sure how much air might leak into the charger’s interior so I smeared some silicone sealant along the joint to minimise air entry so it is now mostly airtight.
Before organising this setup I had ordered from Ebay a USB powered dryer for $12.00
which arrived today. The dryer heats to 45°C (110°F) and It has a timer for three or six hours of drying time.
My question is:
What is the better option, or what combination of desiccant/heating will achieve the best moisture removal from the HAs. Charging for three hours then transferring the aids to the dryer in the middle of the night does not appeal at all.
I have included photos of the charging and drying devices connected to a USB power meter.
Technically inclined Readers may be interested to note that for my slightly discharged Phonak M 70s the charging current was pulsed approx 1 second on at 265milliamps, then 9 seconds off.
The dryer, advertised as operating at 50° C (+ -5°C) In my case the dryer seems to peak at 45° C.
Have you looked in your manual to see what it says about avoiding heat exposure to your HA’s? My ReSound manual for Li-ion rechargeable Quattro’s advises me to avoid heat exposure. Check out what Wikipedia, Battery University, etc., have to say about the effect of heat on the long-term use lifespan of a Li-ion battery. Avoid sub-zero temps, i.e., don’t freeze battery, and best charging temp is about 15 to 25 deg C (55 to 77 deg F). I once totalled a phone battery in short order by always recharging the phone on a very sunny window sill (heat of charging + heat of sun = )!
I checked the manual, and the suggested max is 40 deg C but I take your point any elevation from ambient shortens the ultimate lifespan of Li-ion cells and probably most any cells that function via chemical reactions.
Yes, I imagine any decrease in battery longevity is a time-temperature averaged weighting. Samsung, for example, advises the user in wirelessly charging a phone or watch that the device will get warm but says (for the time involved) that this will not affect the expected lifespan of the device (and, of course, they probably mean you’re supposed to buy a new device from them every two to three years!).
Apple has some of the best advice on Li-ion battery care that I’ve seen anywhere (I guess because their devices are so expensive and they really care about having happy customers).
But in the following link, they mention the optimum temperature for getting the longest lifespan out of their Li-ion batteries - and presumably they’re using top quality stuff. Li-ion batteries can be used at cold temperatures down to about freezing but they should not be charged when very cold - sources vary. I’ve seen a lower limit as low as 45 deg F but I go with 55 deg F as the coldest temperature at which I will charge. Batteries - Maximizing Performance - Apple
" Avoid extreme ambient temperatures.
Your device is designed to perform well in a wide range of ambient temperatures, with 62° to 72° F (16° to 22° C) as the ideal comfort zone. It’s especially important to avoid exposing your device to ambient temperatures higher than 95° F (35° C), which can permanently damage battery capacity. That is, your battery won’t power your device as long on a given charge. Charging the device in high ambient temperatures can damage it further. Software may limit charging above 80% when the recommended battery temperatures are exceeded. Even storing a battery in a hot environment can damage it irreversibly. When using your device in a very cold environment, you may notice a decrease in battery life, but this condition is temporary. Once the battery’s temperature returns to its normal operating range, its performance will return to normal as well."
So, for example, don’t leave your HA’s, your phone, your wireless earbuds, etc., in a car on a hot day. The interior of a car left in a parking lot in southern parts of the U.S. can reach 135 deg F or more on a hot, sunny day in short order.
BTW, for anyone interested in monitoring how hot things are, one can buy very inexpensive thermal “guns” on Amazon in the $15 to $30 price range. You pull the trigger and it measures the infrared radiation being emitted by an object - the more expensive ones have a laser beam “targeting” feature to let you know what you’re “shooting” and adjustments for the infrared emissivity of the type of material, shiny metal surfaces emit less infrared radiation for a given actual temperature than a material that functions more like a “black body.”
Don’t know about the Apple Store but in the Android store you can get battery temperature monitoring apps and set the limit you want the app to beep at if your phone battery gets too hot. The app that I have consumes very little power and shows the temperature read from time to time as a notification as well as flashing a popup and making a noise when I exceed my preset limit. Apple warns you not to charge your iPhone with certain cases on the device. I always remove my phone case when charging my Galaxy Note. I charge my HA’s in their charger case with the lid open. I also remove the case (very carefully!) if I am intensively using the phone and it’s starting to get hot - or I slack off using the phone. Being at home much of the time, I will actually pop my phone into the refrigerator for 5 minutes or so the phone starts to get real hot but I’ve seen advice that temperature cycling, because things are expanding and contracting, is not a good idea to do intensively. I figure the phone normally goes through lots of expansion/contraction cycles just in normal use as its inside gets hot, cools down, gets hot again. So other than moisture condensation, I don’t worry about brief cooling trips to the frig and since my phone has an IP68 rating, I don’t worry about a modest amount of moisture condensation too much, either.
Update: With the thermal gun, you can find the coolest location(s) in your house to store Li-ion battery-powered stuff. If you have central forced-air AC, the coolest locations are usually where the duct work travels the shortest distance from the furnace/blower to a outlet vent - the air warms the least before it’s blown into a room. We’re fortunate enough to have a large walk-in master bedroom closet and I keep my tablets, phone, etc., there when not in use and we’re letting the rest of the house warm to 81 deg F during the summer (or more at times). The closet is at least 5 deg. F cooler than the main part of the house. OTH, running the AC just to keep some battery-powered stuff cool or opening and closing the frig door a lot could cost a lot more in the long run (and waste a lot of your time) than just wearing out a phone by not worrying at all about temperature and getting a new one as needed. I have a lot of battery-powered stuff, including my HA’s, that I’m interested in keeping cool to maximize useful lifespan.
Just curious–so your big closet has it’s own AC vent?
Yes, it does. And my wife does a good imitation of Imelda Marcos, keeping every shoe she’s ever owned in the closet, I think, and taking up most of the closet in the process. We went for a walk the other night and she wore a pair of athletic shoes she’s kept for 30 years or something. They disintegrated during the walk - so much for my theory of temperature preserving things …