I’d be curious to have you wear your headphones and play the ISTS signal through them and record the response with PMM (probe measurement microphones, used during real ear measurements). Or put them on a 2cc coupler in a test box and do it that way. Either way. Then you could do a transition fitting to match the frequency response of the hearing aids toward the frequency response of the headphones. Wouldn’t sound exactly the same due to the difference in the drivers, like you said, but it might bring some of the magic of the headphones to the hearing aids.

That being said, I do think that if you can hear that well with headphones and the cell phone, the biggest improvement would be to allow someone to fit you to REM verified NAL-NL2 targets and have them measure the MPO and keep it under your UCL limits. I would also ask them to run UCL for you across all the same frequencies of a normal hearing test, 250-8k so that it lessens the risk of discomfort from your hyperacusis.

I can tell you are putting a lot of effort into hearing well, and I hope that you can get that improvement.

Hi D Wooluf

Thanks…I have read about them…and agree they look like a possibility. Just don’t have Apple products (anymore)…hoping that the broader market will like them and incentive a wider market.

Ethan

Hi D Wooluf

Agree…how the sound is getting to the aid,vs headphones by wire…quite sure that has an impact on the overall sound envelope.

Ethan

BlockquoteI’d be curious to have you wear your headphones and play the ISTS signal through them and record the response with PMM (probe measurement microphones, used during real ear measurements). Or put them on a 2cc coupler in a test box and do it that way. Either way. Then you could do a transition fitting to match the frequency response of the hearing aids toward the frequency response of the headphones. Wouldn’t sound exactly the same due to the difference in the drivers, like you said, but it might bring some of the magic of the headphones to the hearing aids.

That being said, I do think that if you can hear that well with headphones and the cell phone, the biggest improvement would be to allow someone to fit you to REM verified NAL-NL2 targets and have them measure the MPO and keep it under your UCL limits. I would also ask them to run UCL for you across all the same frequencies of a normal hearing test, 250-8k so that it lessens the risk of discomfort from your hyperacusis.

I can tell you are putting a lot of effort into hearing well, and I hope that you can get that improvement.


Great idea! The Costco exam/fitting was excellent…and followed best practice as far as I can tell. Lots of discussion, repeating tests, final real ear testing, and more tweaking. Left very happy at the time. The worst was the “degreed” Audiologist as truhearing…younger person, with very little technical knowledge as far as I could tell…short test with little discussion. Canceled any further dealings with them. The Costco fitter (not a degreed audiologist) was near retirement, and was the most thorough and knowledgeable of any that I have encountered. Will see him again in two weeks. I will discuss this idea with him.

My working theory is that since the HA use a micro driver directly into the canal, they must loose a lot of overall sound information. Basically the headphones make use of the full ear physiology (ie shape), have an relatively enormous driver, and, very importantly, the muff cancels out most of the world. The end result is that I get as high an original signal to noise ratio as is physically possible.

I have always had a mild tinnitus…and any relatively loud noise would set it off for hours…so between the discomfort of loud sounds in general, and the increase in tinnitus after exposure to loud sounds, I have learned to attenuate and or avoid any loud sounds. Even now with crap hearing.

The most significant benefit I get from (any) HA, is that since world sounds are indeed amplified, my wife can get my attention easier :)…I can’t necessarily hear what she said…but I know I am summoned! Good thing. :slight_smile:

Overall, I like a quiet world. Always have. Loss of hearing sensitivity is definitely a two edged sword for me.

Ethan

1 Like

Thanks! The Costco fitter was concentrating on my hyperacusis. That took up the final third of the fitting appointment. First set the HA to prescription (not sure which standard)…way too loud and VERY uncomfortable…then he did the discomfort test…across all frequencies as noted…and then tweaked the gain to best as he could…did note that the available margin in my lousy left did not leave much room for improvement…but he did the word comprehension test single, both, etc and determined that even with a dialed down left I was getting some comprehension benefit. The overall issue is that this test does not sufficiently mimic real world hearing situations and does not translate into better real-world comprehension in my case. My overall reading of dozens of papers on this subject has validated this point…ie HA in general work far better in the lab than they do in real world. As an engineer, this is of no surprise to me…:slight_smile:
Ethan

This is the only reason my 2 kids survived their teen years…
Dan

3 Likes

In really loud situations i use the Resound Mini Mic. I lay it on the table and it works much better at localization of nearby speech while filtering background noise. You can use Resound devices with Jabra and can buy a mini mic or multi mic on ebay for under $30. I also use the Resound TV streamer that works excellent. I use custom domes and when i changed over from closed power domes it was like night and day. Like molds turbo charged my aids

2 Likes

Thank you! Good information.

Ethan

Yeah I use the multi mic, it’s a fantastic little bit of kit, I use it with the TV as well,saves me having to have extra tv connecters, especially handy when I’m out and about.

2 Likes

Update…a bit new topic, but related.

Still frustrated with the Jabras…so bid on a pair of Phonaks on ebay…were listed at M30’s (direct programmable). Got them yesterday…turned out to be almost brand new P70-R’s! Excellent condition with M receivers (fine for me). Spent an inordinate amount of time programming…starting with the Costo prescription, then used the Phonak Target Direct Audio (a bit of a change)…then tweaked. Bottom line: For 1/5 the price of the Jabras (got an incredible deal on ebay) I am very happy:

  1. Sound Recover 2 settings make a discernable difference (used my wife talking as a benchmark).
  2. Really nice balance between amplification and noise control. Rattling dishes don’t bother me!
  3. Excellent streaming music…far far better than the Jabras…almost as good as my Bose Earbuds.
  4. Excellent phone…I could not really tell where the sound was coming from! Could hear extremely well. Adapted in the truck as I sped up…OS 4?
  5. Bluetooth is a bit sketchy to set up…but now working really well.
  6. Lion batteries in great shape. Charged to 100%…all day with lots of streaming…and plenty of juice left.
  7. Inexpensive option to add a battery to the charger case…$75 new…Jabra/Resound wanted over $200…and I could not actually find to buy…
  8. The Phonak Target software is really easy to use…after lots of study…and seems to be an excellent fitting software…did use the Phonak auto settings…then tweaked the Sound Recover 2 with wife talking. I could discern a very significant difference with it on or off…like going from her talking in a paper bag, to a bit clearer.

I have the 70…so am missing some of the niceties of the 90…but don’t miss yet…any thoughts? I assume these are the same hardware as the 90…anybody have a hack to unlock a 70 to 90?

The Jabras go back next week.

I would appreciate any comments or suggestions.

Ethan

6 Likes

No amount amplification can fix your speech recognition scores…

There are 2 type of hair cells, you need to understand, inner hair cells which is responsible for understanding speech, and outer hair cell which acts like an amplifier of the incoming signal… When the inner hair cells are damaged, you lose speech understanding but when the outer hair cells are damaged, you lose sensitivity to the sound.

I have good speech recognition with moderate to profound loss. Inner hair cells are functional but my outer hair cells are damaged… I hope this makes sense… a young engineer (me) to retired engineer

1 Like

Does the mini mic or multi mic work with Oticon More. If it works, love to buy one.
What is the name of the mini mic or multi mic in Ebay?

Thanks

SSA

Thank you. Makes good sense.

The frequency shifting is somewhat unpredictable I expect because of the physical situation you describe. I hypothesize that the way to handle this is to use a methodology of repeated testing with real world words while iterating through many of the available features of an advanced HA. For example with my P70s, with 20 frequency bins, some control over the frequency shifting (ie the Sound Recover) and other features might result in a modified sound envelope that uses whatever physical recognition I have left. I know how I would do it…I need a HA with some pretty good AI to relatively consistently shift certain word sounds into a envelope that I can actually differentiate…this along with training, should allow a pretty good increase in speech recognition. I already know two things that help…cupping my ear toward the sound…which I expect makes good use of our highly evolved evolutionary ability to sustain damage and lousy hearing environments and recognize a wide variety of sounds dependably, and secondly, concentrating more on the speech to kick in higher order cognitive processing. I want an HA that helps that along…semi intelligently.

One specific way to do that now is to program a semi automatic repetivie test that makes HA changes, tracks changes to word recognition, then uses AI methodology to more quickly optimize available help from the HA. The current methods are too crude…just use one or two tests…then spread out the iteration over days weeks or months. I would speculate that about 30 minutes a day for a couple weeks might be sufficient…the allow that to soak…then repeat in a month or two. This is kinda like what happens now…but far to few test points and iteractions…and the changes may be to subtle and counterintuitive to allow real optimization by us ordinary mortals…

At any rate, by DIY fitting with the P70’s (starting from the “pro” fitting) has already resulted in some perceived improvement…I will see how it goes over the next few months. I will keep repeating the same word recognition test with the same physical controls (volume, location, spatial arrangement, etc) as I make changes over time.

Perhaps someone out there with far better programming ability than I can hack a HA fitting program to include such a testing regime?

Ethan

I found the freqency shifting on oticon hearing aid made things worst in a bimodal situation so i had it completely off. I have a Cochlear Implant on the other ear, like it was causing interference. I used to have it on before i had the surgery it was great to hear things that I could be aware of but that about it. I now have it off because CI provides the full bandwidth of sound of 188 hz to 8khz… My theory is interference because it is putting signal that isn’t what the brain expects so it gets confused…

Update. I just returned the Jabscos to Costco. No issues. Great service. I had a very nice chat with the fitter. He was not aware that the Mic option in the receiver made the Sound Shaper not available. He appreciated the comment and had another high frequency loss customer using the same receiver and may revisit their fit. Very professional response to my comments.

Fitting comment with my new (used P70’s). I tried a number of domes…none seemed to work as well as the ones I got with the Jabscos. I put them on the P70’s…with very good results. Of special note: The M&RIE (Resound/Jabsco) receiver fitting recommends a Surefit 3 “closed” dome. It turns out this dome is different than the other Resound Surefit 3…ie Open, Power and Tulip. The S3 closed dome has two small holes in it…so not totally closed. I found that this specific dome works really well with the P70’s in the better right ear. I use the Resound (Surefit 3) power dome in the left…not much difference than the other power domes I have. I ordered a bunch of the Resound S3 domes.

I see a lot of comments about the important role the domes play…now see that it is likely worth it to try other manufacturers domes with a specific HA.

At this point quite happy the the P70s. I changed the program so a long press gets me mute. From my previous post about HA wish list:

  1. Ear Plug Mode. NRR around 30 db plus or minus. WITH THE MUTE AND POWER IN LEFT AND CLOSED IN RIGHT I GET A DECENT EAR PLUG MODE. NOTE 30DB BY ANY MEANS, BUT USEABLE.
  2. Speech Comprehension mode… Shift s, th, f, t, k to improve comprehension significantly. THE P70 DOES A DECENT JOB…MUCH BETTER THAN I HAD BEEN EXPECTED. I MAXED OUT THE SOUND RECOVER 2 SETTINGS AND GET A DISCERNABLE IMPROVEMENT IN COMPREHENSION.
  3. Music Mode. Like a nice pair of Bose Headphones. Music sounds great with these. I keep the volume pretty low. No signal processing needed. THE P70 IS PRETTY GOOD! NOT UP TO BOSE EARBUDS…BUT PRETTY DANG CLOSE.

Overall the P70s, using mostly the Target calculated fitting, is working pretty well. I admit to being impressed. I actually have been wearing the P70’s all day and forget they are there.

I think Costco will eventually lose HA business if they don’t make up with Phonak.

Thanks so much for all the great comments. Keep them coming.

Ethan

1 Like

Hi Ethan, glad you like the P70’s and managed to program them to your liking, pitty you didn’t end up liking the Jabra in the end, but at least you can call your HAs a hybrid (ReSound Dome, Phonak Body)
As for Costco not selling Phonak, I actually see this as disappointing as well, it would be fantastic if they could make up and sort the issue out, it’s possible but would potentially cause someone to have to swallow a dead rat!

Those that need that feature are outside the M&RIE fitting range from the manufacturer.

1 Like

Good point. But clear that my Costco fitter was not even remotely familiar with that limitation. As a rank, but hard working amateur (afa HA go), I figured it out myself… with lots of help from commenters at this forum! Not any different than what I find in my own (retired) Engineering work…sigh.

Ethan

I would like to emphasize that the Resound Surefit 3 domes fit REALLY well to the Phonak Paradise receivers. They are NOT going to come off without real effort…I see from some other posts that domes coming off when taking out the HA can be a problem. Some generic domes I bought are relative junk…ie slip off the Phonak or Resound Receivers pretty easily.

I would highly recommend the Resound Surefit 3 domes for any receiver that they might fit…the RIE receivers all seem to have a pretty similar mounting geometry (?). I also noted that the Resound S3 domes have a series of small holes rather than just one…appears good for wax infiltration…and I imagine also has an effect on coupled acoustics between the receiver speaker and the ear canal (?).

Ethan