It is a good question, and not one always answered well by the manufacturers even. I’ve seen studies that suggest that hearing aids properly fit to NAL-NL2 outperform ‘major manufacturers’ hearing aids fit to proprietary formulas. The studies almost always obscure what brands and models that they tested, unfortunately. And then the brands have their own studies that usually include a group of like 8 people who did better with their proprietary formula and it never seemed clear to me whether the fitting was verified by any probe mic measurements. I know that isn’t really what you asked, but you should know that this question is also asked by people working in the field. Even though it is a simple question, I think you would be surprised at the variety of answers you would get.
Functional gain provided by the hearing aid can vary widely from person to person due to their anatomy. If it isn’t measured, it could be off by as much as 10dB or more from what is expected. So any formula that is used should be verified. The most common suggestion for ‘best practices’ is probe mic measurements.
Depending on the audiometer used and the fitting software, you can either establish manual targets or some fitting software will include a way that the software can connect to the equipment and it will tell the equipment what targets it is trying to match, basically. In Philips software, there is a tab under the Fit Instruments section that says REM. In that screen, you can select Real Ear Fit. This could allow for a probe mic measured (REM) to be performed with the selected fitting rationale, such as Fit4Speech.
I would like to reiterate that I think NAL-NL2 does a great job in a significant amount of cases, and when it is used by a skilled HIS. If you are hearing well with the hearing aids, I think there is a fair reason to suggest that you should keep those settings. I just wanted to share information since I also think it is good for people to be informed, especially those that are taking their time to seek out that information.
Please be kind to your HIS even if they are not aware of some of this information. I know for a fact there are some members of this forum who have much more knowledge than I do so if they would like to correct anything I’ve said, that is ok too.